• RE: TDS754C fails to boot? no test page on screen

    Thanks for the update. Will locate the 24C02 chips, I probably have some somewhere so will see if I can breath life back in the scope. I would be fine paying for a cal house to cal the unit as I am most fond of this unit.
    Will post an update on findings, thanks Strick.
  • RE: TDS754C fails to boot? no test page on screen

    Tried with a VGA monitor, that produces identical results to the scope display. A momentary screen with horizontal lines and no text.
    The console port revealed the following result, wondering where to start looking.

    DRAM test passed.                 


            Bootrom Header Checksum passed.                                       
            Bootrom Total Checksum passed.                                      
            BootRom Check Sum passed.                                 
            Bus Error Timeout test passed.                                      

    Kernel Diagnostics Complete.                            

    Calling SDM (monitor) Routine.                              

            Enabling Bus Control register. Value = 0x67                                                 
            IMR 1 Register test passed.                                   
            Misc. Register test passed.                                   
            Timer Interrupt test (Auto-Vector) passed.                                                  
            NVRam DSACK test passed.                                
            NVRam Write protected.                              
            Flashrom DSACK and JumpCode test passed.                                                
            Flashrom Checksums passed.                                  

    Bootrom Diagnostics Complete.                             


    DipSwitchValue: 0                 


    Skipping boot loader.                     
    Transferring control to FlashROM.                                 

    No PCMCIA option board detected.                                
    FLOPPY: Detected                

    Adding 6256 symbols for standalone.                                   


    CPU: 68EC040.  Processor #0.                            
    Memory Size: 0x800000.  B                        

    Executing Diagnostics                     
    -> ERRORID: 356 nv storage too small more bytes requested than available                                                                        
    Start Power-On Diag Sequence                            
    hwAccountant probe routines                           
      Probe for unexpected pending ints                                   
      Dsp Instr mem size                    
      Dsp D2 mem size                 
      Dsp D1 mem size                 
      Dsy Vect0 mem size                    
      Dsy Vect1 mem size                    
      Dsy Wfm0 mem size                   
      Dsy Wfm1 mem size                   
      Dsy Text0 mem size                    
      Dsy Text1 mem size                    
      Acq number of digitizers                          
      Acq mem size              
      Cpu timer interval uSec                         
      Cpu Dram size               
      NvRam mem size                
      Limit to 1GS/sec presence                           
      Opt Math Package prese                      
      Opt Telecom Masks presence                            
      Opt 3C presence                 
      Opt 4C presence                 
      Opt RS232/ Cent presence                          
      Opt 1M presence                 
      Opt 2M presence                 
      Acq Intlv Cal Id presence                           
      Opt TvTrig presence                     
      Opt TvTrig index                  
      Dsy color presence                    
      Opt floppy drive presence                           
      Opt hard drive presence                         
      Acq number of user channels                             
    dspForcedBus ................... pass                                     
    cpuDiagD2MiscReg ............... pass                                     
    cpuDiagDSPIntMaskReg ........... pass                                     
    cpuDiagDsyAccess ............... pass                                     
    dsp68kMemTest .................. ***FAIL***                                           
    ..error details:                
    ERRORID: 162 diagnostic test failure dsp68kInMemTest                                                    
    ** addr = 0x7000000  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000004  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000008  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x700000c  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000010  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000014  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000018  exp data = 0x55555500  actual = 0x5555ff00                                                               

    calLibrarianDefaultCk .......... pass                                     
    dspForcedBus ................... pass                                     
    acqProcThermistor .............. pass                                     
    trigGtlRegisterDiag ............ pass                                     
    trigBtlRegisterDiag ............ pass                                     
    ch1EdgeTrigDiag ................ pass                                     
    lineTrigDiag ................... pass                                     
    dlyTrigDBTRunsAfter ............ pass                                     
    slewrateTrigDiag ............... pass                                     
    trigAttenSerialReg ............. pass                                     
    trigPreampSerialReg ............ pass                                     
    trigDTCSerialReg ............... pass                                     
    trigExtlSerialReg .............. pass                                     
    trigDacSerialReg ........                       
    TICountersDiag ................. pass                                     
    gtlBigCountersDiag ............. pass                                     
    trigBtlConfidenceDiag .......... pass                                     
    trigGtlCompRamDiag ............. pass                                     
    digRegisterConf ................ pass                                     
    digSpecialRegisterConf ......... pass                                     
    fpDiagConf ..................... pass                                     
    optDiagPM110Reg ................ pass                                     
    optDiagFloppyCacheMem .......... pass                                     
    optDiagFloppyControllerIO ......0x5fffe80 (tRootTask): libError 356, lib EXTCAL,                                                                                
     id=356, msg=more bytes requested than available                                                
     pass     
    optDiagFloppyDrive ............. pass                                     
    optRS232DuartIO ................ UNTESTED                                         
    optRS232DuartIntLoop ........... UNTESTED                                         
    optCentronCntrlReg ............. UNTESTED                                         
    optTv8bitRegDiag ............... pass                                     
    optTvShiftRegDiag .............. pass                                     
    optTvXparentCodes .............. pass                                     
    optTvDontCareCodes ............. pass                                     
    optTvEdgeAndLevel .............. ***FAIL***                                           
    ..error details:                
    ERRORID: 162 diagnostic test failure optTvEdgeAndLevel                                                      
    +edge SWT0-SIG1, no NTRIG2 output                                 
    +edge SWT1-SIG1, no NTRIG2 output                                 
    +edge SWT0-SIG2, no NTRIG2 output                                 
    +edge SWT1-SIG2, no NTR                     

    Executing Smalltalk                   
    ERRORID: 356 nv storage too small more bytes requested than available                                                                     
    hwAccountant probe routines                           
      Probe for unexpected pending ints                                   
      Dsp Instr mem size                    
      Dsp D2 mem size                 
      Dsp D1 mem size                 
      Dsy Vect0 mem size                    
      Dsy Vect1 mem size                    
      Dsy Wfm0 mem size                   
      Dsy Wfm1 mem size                   
      Dsy Text0 mem size                    
      Dsy Text1 mem size                    
      Acq number of digitizers                          
      Acq mem size              
      Cpu timer interval uSec                         
      Cpu Dram size               
      NvRam mem size                
      Limit to 1GS/sec presence
      Opt Math Package presence
      Opt Telecom Masks presence
      Opt 3C presence
      Opt 4C presence
      Opt RS232/ Cent presence
      Opt 1M presence
      Opt 2M presence
      Acq Intlv Cal Id presence
      Opt TvTrig presence
      Opt TvTrig index
      Dsy color presence
      Opt floppy drive presence
      Opt hard drive presence
      Acq number of user channels
    0x5fffe80 (tRootTask): libError 356, lib EXTCAL, id=356, msg=more bytes requeste
    d than available
    can't open input 'fd0:/startup.bat'
      errno = 0x13 (S_errno_ENODEV)

    Smalltalk/V Sun Version 1.12
    Copyright (C) 1990 Object Technology International Inc.
    ERRORID: 162 diagnostic test failure extended cal librarian reset
     
  • TDS754C fails to boot? no test page on screen

    My TDS754C having stood unused a while while moving home, now will not boot up.
    A blue pattern appears on screen and all buttons light, relays click, then lights go out but no boot.
    Checked nvr battery voltage as good.
    Help needed for where to start checking fault please,
    Recorded the following from console port on booting:

    DRAM test passed.                 

            Bootrom Header Checksum passed.                                       
            Bootrom Total Checksum passed.                                      
            BootRom Check Sum passed.                                 
            Bus Error Timeout test passed.                                      

    Kernel Diagnostics Complete.                            

    Calling SDM (monitor) Routine.                              

            Enabling Bus Control register. Value = 0x67                                                 
            IMR 1 Register test passed.                                   
            Misc. Register test passed.                                   
            Timer Interrupt test (Auto-Vector) passed.                                                  
            NVRam DSACK test passed.                                
            NVRam Write protected.                              
            Flashrom DSACK and JumpCode test passed.                                                
            Flashrom Checksums passed.                                  

    Bootrom Diagnostics Complete.                             


    DipSwitchValue: 0                 


    Skipping boot loader.                     
    Transferring control to FlashROM.                                 

    No PCMCIA option board detected.                                
    FLOPPY: Detected                

    Adding 6256 symbols for standalone.                                   


    CPU: 68EC040.  Processor #0.                            
    Memory Size: 0x800000.  B                        

    Executing Diagnostics                     
    -> ERRORID: 356 nv storage too small more bytes requested than available                                                                        
    Start Power-On Diag Sequence                            
    hwAccountant probe routines                           
      Probe for unexpected pending ints                                   
      Dsp Instr mem size                    
      Dsp D2 mem size                 
      Dsp D1 mem size                 
      Dsy Vect0 mem size                    
      Dsy Vect1 mem size                    
      Dsy Wfm0 mem size                   
      Dsy Wfm1 mem size                   
      Dsy Text0 mem size                    
      Dsy Text1 mem size                    
      Acq number of digitizers                          
      Acq mem size              
      Cpu timer interval uSec                         
      Cpu Dram size               
      NvRam mem size                
      Limit to 1GS/sec presence                           
      Opt Math Package prese                      
      Opt Telecom Masks presence                            
      Opt 3C presence                 
      Opt 4C presence                 
      Opt RS232/ Cent presence                          
      Opt 1M presence                 
      Opt 2M presence                 
      Acq Intlv Cal Id presence                           
      Opt TvTrig presence                     
      Opt TvTrig index                  
      Dsy color presence                    
      Opt floppy drive presence                           
      Opt hard drive presence                         
      Acq number of user channels                             
    dspForcedBus ................... pass                                     
    cpuDiagD2MiscReg ............... pass                                     
    cpuDiagDSPIntMaskReg ........... pass                                     
    cpuDiagDsyAccess ............... pass                                     
    dsp68kMemTest .................. ***FAIL***                                           
    ..error details:                
    ERRORID: 162 diagnostic test failure dsp68kInMemTest                                                    
    ** addr = 0x7000000  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000004  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000008  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x700000c  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000010  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000014  exp data = 0x55555500  actual = 0x5555ff00                                                               
    ** addr = 0x7000018  exp data = 0x55555500  actual = 0x5555ff00                                                               

    calLibrarianDefaultCk .......... pass                                     
    dspForcedBus ................... pass                                     
    acqProcThermistor .............. pass                                     
    trigGtlRegisterDiag ............ pass                                     
    trigBtlRegisterDiag ............ pass                                     
    ch1EdgeTrigDiag ................ pass                                     
    lineTrigDiag ................... pass                                     
    dlyTrigDBTRunsAfter ............ pass                                     
    slewrateTrigDiag ............... pass                                     
    trigAttenSerialReg ............. pass                                     
    trigPreampSerialReg ............ pass                                     
    trigDTCSerialReg ............... pass                                     
    trigExtlSerialReg .............. pass                                     
    trigDacSerialReg ........                       
    TICountersDiag ................. pass                                     
    gtlBigCountersDiag ............. pass                                     
    trigBtlConfidenceDiag .......... pass                                     
    trigGtlCompRamDiag ............. pass                                     
    digRegisterConf ................ pass                                     
    digSpecialRegisterConf ......... pass                                     
    fpDiagConf ..................... pass                                     
    optDiagPM110Reg ................ pass                                     
    optDiagFloppyCacheMem .......... pass                                     
    optDiagFloppyControllerIO ......0x5fffe80 (tRootTask): libError 356, lib EXTCAL,                                                                                
     id=356, msg=more bytes requested than available                                                
     pass     
    optDiagFloppyDrive ............. pass                                     
    optRS232DuartIO ................ UNTESTED                                         
    optRS232DuartIntLoop ........... UNTESTED                                         
    optCentronCntrlReg ............. UNTESTED                                         
    optTv8bitRegDiag ............... pass                                     
    optTvShiftRegDiag .............. pass                                     
    optTvXparentCodes .............. pass                                     
    optTvDontCareCodes ............. pass                                     
    optTvEdgeAndLevel .............. ***FAIL***                                           
    ..error details:                
    ERRORID: 162 diagnostic test failure optTvEdgeAndLevel                                                      
    +edge SWT0-SIG1, no NTRIG2 output                                 
    +edge SWT1-SIG1, no NTRIG2 output                                 
    +edge SWT0-SIG2, no NTRIG2 output                                 
    +edge SWT1-SIG2, no NTR                     

    Executing Smalltalk                   
    ERRORID: 356 nv storage too small more bytes requested than available                                                                     
    hwAccountant probe routines                           
      Probe for unexpected pending ints                                   
      Dsp Instr mem size                    
      Dsp D2 mem size                 
      Dsp D1 mem size                 
      Dsy Vect0 mem size                    
      Dsy Vect1 mem size                    
      Dsy Wfm0 mem size                   
      Dsy Wfm1 mem size                   
      Dsy Text0 mem size                    
      Dsy Text1 mem size                    
      Acq number of digitizers                          
      Acq mem size              
      Cpu timer interval uSec                         
      Cpu Dram size               
      NvRam mem size                
      Limit to 1GS/sec presence
      Opt Math Package presence
      Opt Telecom Masks presence
      Opt 3C presence
      Opt 4C presence
      Opt RS232/ Cent presence
      Opt 1M presence
      Opt 2M presence
      Acq Intlv Cal Id presence
      Opt TvTrig presence
      Opt TvTrig index
      Dsy color presence
      Opt floppy drive presence
      Opt hard drive presence
      Acq number of user channels
    0x5fffe80 (tRootTask): libError 356, lib EXTCAL, id=356, msg=more bytes requeste
    d than available
    can't open input 'fd0:/startup.bat'
      errno = 0x13 (S_errno_ENODEV)

    Smalltalk/V Sun Version 1.12
    Copyright (C) 1990 Object Technology International Inc.
    ERRORID: 162 diagnostic test failure extended cal librarian reset
     
  • TDS754C fails to boot? no test page on screen

    Hi, I have a scope that sat off mains for some time while i moved house. Now it fails to boot up, there is a flash of blue screen for a second then you hear the relays and the panel button lights go out, the LED display on the A board flashes top & bottom after a number sequence. I checked the NV chips battery voltage as I suspected a fail but both have good battery voltage. Sw8 open then shows the led display go thru a number sequence and still no boot up. Looking for guide as to where to begin looking.