Setting sample rate on DPO4104 Oscilloscope
I am writing a python script that interacts with a DPO4104 Oscilliscope using SCPI over eternet via pivisa. I have gotten many features to work following the commands in the DPO4000 Series programmer manual. I am stuck trying to set the sample frequency.
I am able to read the frequency with the command:
horizontal:samplerate?
but I am unable to change the command with:
horizontal:samplerate 100000I have tried the abbreviated form of the command, and tried decimal and scientific notation for the value. Any attempt does not seem to have any effect on the oscilloscope, and I am not getting any errors in python. Below is minimal code to reproduce (hopefully) the problem I am having.
import pyvisa
rm = pyvisa.ResourceManager()
scope = rm.open_resource("TCPIP::192.168.20.82::INSTR")
print(scope.query("*idn?"))
print(scope.query("horizontal:samplerate?"))
scope.write("horizontal:samplerate 500.0000E+3")
print(scope.query("horizontal:samplerate?"))Output:TEKTRONIX,DPO4104,B010495,CF:91.1CT FV:v2.481.0000E+61.0000E+6