RE: Missing Trash Can on MSO56 screen
Hello,
I believe that you are referring to the recycling bin on the desktop of a Windows 10 based scope. If that is not correct, could you please elaborate?
If your scope is running Windows 10, you can show the recycling bin with the following steps.
Open the Start menu > select Settings.
Go to Personalization > Themes > Desktop Icon Settings
Select RecycleBin > Apply
-
RE: SCPI/programming: Is there a way to simply retrieve the scope screen shot as a file object
Hello,
One method for saving the screen capture directly to the host computer is to create a shared folder on the host computer and connect to it from the scope. The scope could then save data to the shared folder, this would also work for saving waveforms, and session files. -
RE: AFG3000 series and Arbexpress - creating a sine waveform
For setup B you could use the Modulation Mode set to AM, with 100% modulation depth, and internal modulation source. Create a custom waveform in ArbExpress of just the pulses scaled to ±1V. Values of -1V will be off, and +1V full amplitude. -
RE: MSO5 series spurious tone
Hello,
Spurs at integer fractions of the sample clock are normal. These can be slightly reduced by running regularly Signal Path Compensation (SPC) after the scope has warmed up. For low level measurements we recommend running SPC once a week, or if the ambient temperature has changed by more than 5C. You can run SPC by pressing Utility > Calibration > Run SPC. -
RE: RSA306B What is the physical setup required to measure SINAD?
In addition to Afonso's comment, you will need to connect the signal to the RF input (type N connector) on the RSA306B. The Ref In port is for connecting an external 10MHz refrence clock if desired. -
RE: MSO58 LP CAD Model
Hello,
Attached you will find the MSO58LP CAD model. -
RE: how to get EXIT key file in kickstart for checking it in AMS
Hello,
To remove a license from KickStart and to generate the exit key do the following:- Click on the key icon in the top right corner of the KickStart application.
- In the pop up window select the license you would like to remove.
- Click "Remove License"
- Save the exit key somewhere easy to find on your system.
- For ease I like saving it to a folder in Documents
-
RE: MSO64 - TPP0200/1 and Probe Tip to BNC P/N
The TPP0200 will work with the MSO64, it has a compensation range of 8pF to 18pF which works with the MSO64's 14.5pF input capacitance. While the TPP0201 has a compensation range of 15pF to 25pF, thus it cannot be compensated to the MSO64.
The part number for the BNC to probe tip adapter is 013-0367-xx. -
RE: Latest firmware for the TDS3000/ TDS3000B series?
Firmware version 3.41 for the TDS3k and TDS3kBs can be found at the link below.
TDS3000 AND TDS3000B SERIES FIRMWARE FV3_41 | Tektronix -
RE: UI Bug in V1.40 Firmware for MSO 5-Series B Scopes
This bug applied to MSO4/5/6 scopes and has been resolved with V1.44.3 firmware released yesterday 10/13/2022.