RE: Methods to Improve the Data Transfer Speed and Increase the Update Rate of AWG5204
What is the size of Waveforms (in terms of sample) that you need to transfer remotely to the AWG?
the longer the waveform, the longer is the time to transfer and play the waveform out.
Have you considered the Sequencer Option for your Application, so you don't have to transfer all waveform from a PC?
My recommendation is reaching our to a Local Application Engineer in your region. -
RE: Spectrum analyzer RSA306B "unstable test data" at no input signal
Hi Jeremy,
The step jump around 20 MHz is due to differences in noise figure of separate signal paths used in the architecture’s frequency plan. This is not an uncommon artifact of spectrum analyzers when sweeping across wide frequency ranges.
That said, the measurement accuracy remains to our published specifications and we do not believe this should impact your ability to perform EMI measurements.
-
RE: RSA5103A The software could not be enabled
This is hardware failure and will require Repair, please contact your Local Tektronix Service. -
RE: Burst measurement channel 3
please contact Technical Support, depending on the Region: Support | Tektronix -
RE: Burst measurement channel 3
I would recommend contacting your Local Tektronix Sales Representative and the Regional Technical Support for this topic.
Here I can only point to the Datasheet, frequency Burst is indicated for channel A and B: Microwave/Counter/Analyzer with Integrated Power Meter | Tektronix
-
RE: Burst measurement channel 3
Frequency Burst is not available on Input C of MCA Models. This feature was removed.
It is available on FCA3020 and FCA3120 – 20 GHz Only