RE: Which probes work with the MDO 2 series
These three probes are compatible with the 3 series MDO. The TDP1500 does have a higher bandwidth than our 3 series, but this will not affect the compatibility up to the scope bandwidth limit.
-
RE: Add CH1 and CH2 signals on an AFT 1062
I recommend using a 50 ohm resistive power splitter/combiner like the Mini-Circuits ZFRSC-2050+ (https://www.minicircuits.com/WebStore/dashboard.html?model=ZFRSC-2050%2B). Using a BNC-tee won’t damage your AFG but you might not get the amplitude you expect due to double loading. What amplitude (and frequency) are you looking for and is this something you need often?
-
RE: Can I receive an output from Signal Express without the program on my FCA 3100?
Yes, you can get information to a computer without SignalExpress by writing some software. Please see the programmers manual below for examples and commands to control the FCA3000.
https://www.tek.com/en/frequency-counter/mca3000-manual/fca3000-and-fca3100-series-mca3000-series
-
RE: Remote PC Access a DPO 2024
You can remotely view the screen and change settings with e*Scope. To use e*Scope, connect your scope to ethernet and then enter the IP address into your browser address bar. You will need to make sure that your computer and your scope are on the same network.
Need DPO2CONN for ethernet
-
RE: Verification/Calibration/Manual for the 2110
See the link below for the 2110 Service manual with the PV and adjust information. https://www.tek.com/en/manual/digital-multimeter/broad-purpose-digital-multimeters-manual-3-keithley-2110-series-5-digit-dual-display-usb-multimeter
Primary Manual can also be found here:
https://www.tek.com/en/manual/digital-multimeter/broad-purpose-digital-multimeters-manual-3-keithley-2110-series-5-digit-dual-display-usb-multimeterOther links related to the 2110:
https://www.tek.com/en/search?keywords=2110&facets=_templatename%3dManual&sort=desc -
RE: TDS 1002 PC Software
The TDS1000 required an option communications module to communicate to a PC. All of these modules are no longer available for sale. If you have a module for the scope please reference the programmers manual for additional details.
Modules:
TDS2CM or TDS2CMA, any version, when used in TDS1000 or TDS2000 Series instruments, any version.TDS2MEM any version, when used in most TDS1000 or TDS2000 Series instruments (except TDS1001 and TDS2004 models), any version.
-
RE: SDLC serial protocol support on the MDO 4104C
SDLC is suppoted on MSO4, MSO5B, MSO6B, not supported on MDO4000
-
RE: User manual for TDS2024B
You can find the user manual here: https://www.tek.com/en/oscilloscope/tds1001b-manual/tds1000b-and-tds2000b-series
-
RE: Product serial numbers and information on older scope accessories (i.e. 453A)
The best place is this customer's website: https://w140.com/tekwiki/wiki/453
Another one would be VintageTEK:
https://vintagetek.org/Serial number location could be right under the 543A-4 on front.
Or lower left by Channel one on the 453A-1.
Sorry, not sure for the Rackmount verison.
Very old scopes liek this were usually on the front.
1980's and up were on the back. -
RE: Digital Probe for the MSO 4054
The MSO4000 Series oscilloscope uses the P6516 logic probe, part number 071214901.