RE: Device characterization tests with 4200A-SCS and Lakeshore probe station
Hi Andrea, thanks for your response. Yes, I am from the University of Nebraska-Lincoln. Can you please provide the contact address for local Keithley FAE? -
Device characterization tests with 4200A-SCS and Lakeshore probe station
We have a 4200A-SCS connected with a lakeshore probe station, where we can test our SiC/GaN dies through Clarius software. There we can perform I-V test, threshold voltage etc. in low voltage (-210 to 210 V) and low current setting (-0.105 to 0.105 A). Now for some new device testing, we need to increase the voltage and current setting (up to 650 V, 50 A). Can we integrate the 2657A (HV SMU) and 2651 A (HC SMU) with the existing 4200A for probe station testing? -
RE: Using 2657A for Double Pulse Test
Thank you for your response. Can we use Model HV-CS-1613 and directly use it in the DPT PCB board? -
Using 2657A for Double Pulse Test
For DPT, we are planning to provide the drain bus voltage by SMU 2657A. That 2657A is connected with 8010A with the <strong>HV-CA-554</strong> Triax cables. To use those cables in our DPT board's BNC input, we tried <b>237-BNC-TRX</b> from Keithley but unfortunately <b>it doesn't match with the HV-CA-554 cable</b>. Can you please suggest any <b>TRX to BNC adapter </b>that can match with these cables?