RE: TBS1052C multiply problem
Afonso - I must say this is very poor service. We bought around 50 oscilloscopes from you that are less than 12-months old that have a serious, fundamental fault. I informed Tektronix of the problem over a month ago. How many users are out there that are using the multiply function without realising it is in error - the problem was obvious for the simple set up we have.
We renew our oscilloscopes on a 3-year basis - we won't be using Tektronix again.
Andrew
-
RE: TBS1052C multiply problem
Thanks Afonso - I'll try your suggested workaround.
Andrew -
TBS1052C multiply problem
Our set of TBS1052C oscilloscopes exhibit a strange effect when multiplying two waveforms. The attached shows the multiplication of two sine waves (yellow and blue waveforms) with equal frequency, 2:1 amplitude and a small phase difference. The first screenshot (Waveform1.png) shows the expected result from multiplication (red waveform) - a 2x frequency sinewave with DC offset. However, when changing the volts/per division of one channel, the second multiplied waveform becomes distorted (Waveform1.png). We have tried different oscilloscopes of the same model and they all exhibit the same problem. However, interestingly the version B doesn't have the problem.