RE: Probe quantity when ordered
The TBS2074 should include 4 probes.
Models
TBS2074 70 MHz, 1 GS/s, 20 M record length, 4-channel digital storage oscilloscopeStandard accessories
Probes TPP0100 100 MHz, 10x passive probe (one per analog channel) -
RE: Probe Compatibility
The TCP0030A is compatible with the DPO7054. Feel free to reach out if you have any other questions, if you'd like to order this probe you can email buy@tektronix.com
-
RE: Probe adapter to interface a P5205A probe to a DPO2024 or a MSO44
The TPA-BNC adapter would be required to use the P5205A with both the DPO2024 and the MSO44. Unfortunately the DPO2024 cannot supply enough power to operate the P5205A so it would also require the 1103 probe power supply.
Let me know if you have any further questions that I can help you out with.
-
RE: Pre-compliance testing and PDN measurments
Hello,
The DPO5104B comes with FFT as a base feature, you can find information on it starting on page 132 of the user manual. The DPO5104B is a real time scope, it is not a sampling scope.
If you would like a more automated way to conduct pre-compliance I would suggest using one of our RSAs and EMCVu software.
MSO/DPO5000B User Manual:
https://www.tek.com/en/oscilloscope/dpo70000-mso70000-manual/mso70000dx-dpo70000dx-mso70000c-dpo70000c-dpo7000c-mso5000b-0 -
RE: North American power cord for the MSO22 A0 and MSO24 AO
Thanks for contacting us,
The North American power cord is option A0 and can be ordered as MSO22 A0 or MSO24 A0 (The cord's standalone part number is 161034800). Just for clarity, this cord should not be confused with the AC/DC power adapter for the scope which it plugs into (The power adapter is part no. 119972500). -
RE: Calibration requirement for MSO2024B
We recommend every 12 months for calibration. It looks like our service group does 17025A and 17025C calibration levels.
If you are interested in calibration please reach out to our service team.
Below you can find the contact information for our Service Department and someone will be happy to help:
Service and calibration information.
800-438-8165
Web: https://www.tek.com/rma-request
Email: rmarequest@tektronix.com -
RE: How to play back r3f files saved from the RSA306B
Hello,
To enable playback .r3f files you will need to move the license SV56FL-SVPC off the RSA and on to your computer. I would also recommend downgrading to V4.1.0022 of SignalVu-PC as it is the latest version of SignalVu validated for the RSA306B.
To move the license from the RSA to SignalVu generate the exit key from the RSA, check it back into TekAMS, and check it out to the computer without the RSA connected. This will allow you to playback recorded sessions.
-
RE: TBS1000 Series XY mode
Hello,
It looks like your Ch 2 has been setup to measure current. To change this back to voltage, push the Ch 2 menu button and select Probe Setup. Use the multipurpose knob to hover over Measure Current and push the knob to toggle back to No. This should change the scale back to volts and fix the 10x amplitude problem. When the scope is setup to measure current, it auto sets the volts to amps conversion as 10A/V.
-
RE: Would Like To See TBS1052B Data On PC.
Hello,
There are a couple different ways you can transfer the waveform data over to your computer.
1) USB flashdrive to save a .csv file and manually transfer over to your PC
2) PC software, such as TekScope Utility: https://forum.tek.com/viewtopic.php?t=140451
3) Programmatically pull the waveform data using SCPI commands -
RE: DAQ6510 won't detect scan card
Based on the information provided, it looks like this needs to be repaired. For a RMA repair quote contact:
Tektronix/Keithley Repair/Calibration:
Web: https://www.tek.com/rma-request
Email: rmarequest@tektronix.com