Is this a bug in MSO64B during measurement badge readings for Maximum and Minimum
In the eScope screen capture image attached (Top MSO64B_Max_Min_Reliability.jpg): the Maximum and Minimum badge values report 268.1 mV and -272.0 mV respectively.
But just from eyeball measurement the Maximum and Minimum values should be around 2.7 V and -2.7 V respectively.
After turning off the scope, and restarting. The readings seem correct as seen in this image: Bottom MSO64B_Max_Min_Reliability.
I see there is a new firmware verison, so I will upgrade. But perhaps you folks could like into what makes these readings unreliable.
Is my scope flaky or is this a bug in scope firmware?
Scope credentialsManufacturer Description Tektronix Oscilloscope MSO64B C031954 Firmware Version 2.8.1.1496 -
RE: Customer documentation survey
Error 1: <br> I just noticed that a very basic command is missing from 4-5-6-series-mso-programmer-manual ID:  <p><i>077-1305-23</i></p> <br> MEASUrement:MEAS<nr>:VALUE?<br> <br> While these commands are documented: <br> <br> MEASUrement:MEAS<nr>:YUNIT?;TYPE?<br> <br> Please review programmer manual and make sure all MEASUrement:MEAS<nr> options are documented.<br> <br> Error 2:<br> <br> page 3-8 Table 3-3: instrument operations that can generate OPC<br> <br> This command  <p><i>MEASUrement:MEAS<x> :RESUlts<br> <br> Has a space between > and : when there should be no space.</i><br> The result is that I copy this command to search for more details in the programmer manual but it is not found anywhere. But remove the space - and the command is found throughout the document.<br> This detail cost me hours of delay in learning how to program my MSO64B scope.</p> <br>  