-
RE: Unable to read data from Tektronix 5034 Oscilloscope in LabView 2022
Hi,
Correct me if I'm wrong, but I presume you are using a DPO/MSO/MDO5034 and not the TDS5034.
A fresh install of Windows might be worth considering if using LabView is important to you.
I would suggest trying TekScope Utility. It should allow you to easily save waveform data manually or on trigger or timing events.
If it does not do what you need it to do, then you may want to try sending SCPI commands manually through OpenChoice and see if the data you receive matches with what you'd expect. Then it's a simple matter of stringing those together into Python or some other language of your choice.
If you post your LabView program we might be able to provide more direct help on it.
I would also point to this post that features several known working LabView remote control examples. -
RE: Is it possible to transfer both analog waveform and spectrum view data in one trigger event?
Hi Jeremy,
Just to be clear, the data is already in log form if using dB units. If using Vrms then the data will be linear.
The data follows the units that spectrum view is set to.
"DAT:WIDTH 8" Is actually not necessary as when we set the data:source to CHx_SV_NORMAL, it automatically sets the data width to 8. This is because the spectrum view data is encoded as double precision floating point, which takes up 8 bytes per point, as mentioned on page 2-420 of the programmer's manual.
So it's there just to make it clear to any reader that the data is 8 bytes per point.
-
RE: Is it possible to transfer both analog waveform and spectrum view data in one trigger event?
2. That is correct, the number of data points available is equals to the record length that the instrument is configured to.
3. I have attached a Python script I wrote that exemplifies how to read the spectrum waveform data from the scope. This can be a bit tricky as it is a double float and needs to be converted into a usable format from its packed binary representation.
To your initial question, no you should not apply a log function to this data as it is already in log form. From page 2-429 of the manual:
> CHx_SV_NORMal, CHx_SV_AVErage, CHx_SV_MAXHold, CHx_SVMINHold: When DATa:SOUrce is set to one of these traces, the data points represent the amplitude of the Spectrum View frequency domain traces in the current units, and are floating point values -
RE: Is it possible to transfer both analog waveform and spectrum view data in one trigger event?
1. Yes, it is possible to transfer both the time domain waveform and the spectrum view data in one trigger event. This might require stopping acquisition to allow time to transfer both waveforms.
The argument for DATA:SOURce should be "CHX_SV_NORMAL", as specified on page 2-439 of the programmer's manual for the MSO4.
How are you processing the data? It should be floating point data (as mentioned on page 2-429), so they take a bit more work to read than time domain waveform data.
2. That is up to how many points you wish to read. If you wish to read all the points then you can send "DATA INIT" to set it to the default settings.
3. I believe this issue goes back to I discussed in 1., that the data is floating point and so needs to be read differently from time domain waveform data. -
RE: Difference between RF function on 3 Series MDO and 4 Series MSO
Hi Jeremy,
They are significantly different. The following comparison explains it well: <a href="https://www.tek.com/en/documents/fact-sheet/3-series-mdo-spectrum-analyzer-vs-oscilloscope-fft-comparison">https://www.tek.com/en/documents/fact-sheet/3-series-mdo-spectrum-analyzer-vs-oscilloscope-fft-comparison</a>
Main points to take away:- ~15 dB better dynamic range than scopeFFT
- Significantly higher bandwidth (Up to 3 GHZ)
These may, or may not, be important to you, depending on what measurements you are taking.
-
RE: Save large CSV files of Signal Waveform (Voltage Measurement)
One hour is quite a long acquisition for a scope. This is the kind of measurement that is usually best done by a digital multimeter.
There is also a question of the sample rate you need this acquisition to be done at, which you did not specify.
I would suggest using TekScope Utility (https://forum.tek.com/viewtopic.php?t=140451) and using its Data Logger feature to record several waveforms that add up to one hour of acquisitions. -
RE: Save full-resolution image from MDO34 (3 Series) via network
Hi Jacob,
Yes, the option to change the default save location is under Tools -> Options -> General -> Default Save Locations. -
RE: Save full-resolution image from MDO34 (3 Series) via network
Hi Jacob,
There are several ways to connect your PC to your MDO34. eScope is meant as a quick and easy way to remote control the instrument.
If all you need to do is get a screenshot/hardcopy, I would recommend the free application, TekScope Utility.
It also features other data acquisition capabilities, like waveform data or measurement values.
For something more fully fledged, TekScope is the way to go. It allows you to connect to and control several scopes at the same time, controlling them all through the same interface as you will be used on the instrument itself. -
RE: Saving and exporting multiple waveforms in quick succession
Hi David,
The SAVE commands are generally used for saving to a mounted drive.
It appears what you want is to transfer the waveform over the connection.
To do this, you should use the following sequence:
"CURVE?" outputs the waveform data and the following line reads that into an array you can then process and/or save as you wish.scope.write('CURVE?') data = scope.read_raw()
-
RE: How to select type 'Time out' for MDO34 programmatically
Hi Jeremy,
Indeed the SCPI interface for the 4 and 3 series are slightly different.
The programmer's manual describes how to setup the timeout trigger on page 2-67 (picture attached). You will notice it is a different procedure from that you expected from the MSO4.