RE: AFG1022 does not properly load arb waveforms
I'm unable to replicate your issue, would you be able to post the definitions of wvfm_setlen() and q_error()?
The raw SCPI output from that section would also work. -
RE: AFG1022 does not properly load arb waveforms
I've reported that issue and I'll be working with the software team to get it fixed.
A power cycle isn't enough to reset the AFG to default state, instead you should issue the
*RST
command to default set the instrument and then proceed with your tests. Let me know if that still results in the same error or not! -
RE: How to query AFG1022 if it has finished processing commands?
Hi Boris,
Yep your test is great and I've also ran similar tests which confirms that
*OPC?
works to synchronize commands. Specifically, running*OPC?
by itself takes 35 ms, doingdata:copy
is about 1 ms, but if we do both it takes 50 ms, which implies that*OPC?
is waiting about 15 ms fordata:copy
to finish, as expected.I also tested using other queries like
system:err?
ordata:value?
and those had the same effect, which lead me to believe the AFG's programming interface is synchronous, so explicit synchronization with*OPC?
is likely not needed. -
RE: How to query AFG1022 if it has finished processing commands?
Hi Boris,
Probably the best explanation of *OPC? is in this programmer's manual: https://www.tek.com/en/sitewide-content/manuals/4/5/6/4-5-6-series-mso-programmer-manual
Essentially, you'd wait until you read 1 from *OPC? and that would tell you that your OPC operations are done. I've attempted to try it out on the AFG1k but haven't been able to confirm it works.
The *WAI command might also be of interest to you.
Also, a lot of our signal sources are synchronous, making it unnecessary to query *OPC?. I'm still trying to confirm if this is the case for the AFG1k or not.
Do you have a specific setup which has been giving you synchronization problems?
-
RE: How to query AFG1022 if it has finished processing commands?
Hi Boris,
It's not documented, but the instrument does reply to an *OPC? query. I haven't tested it thoroughly on this instrument, but does how one would usually do synchronization through SCPI. -
RE: AFG1022 does not properly load arb waveforms
Hi Boris,
I'm not quite sure why you're getting that -201 error, I would assume it has something to do with the commands you sent previously to the AFG which set its state, but I've been unable to replicate it, so I can't explain it. If you could share more of your script/commands I might be able to track it down further.
In terms of the waveform upload issue with 500 pts, I was able to replicate that and I've tracked it down to a problem that happens with waveforms that are roughly a multiple of 250 pts. So the current workaround is just to avoid those lengths, but I will be working with our teams to fix this issue.
-
RE: AFG1022 does not properly load arb waveforms
Hi Boris,
In my script I usedsource1:function EMEMORY
(no :shape) and that seems to work.
I've attached my python script with this post for further reference.
-
RE: How to generate chirp (sweep signal) in AFG31000?
Hi Avital,
In addition to what Daniel said, you can also:
1) If your chirp width is 2 MHz or less, you can use the Modulation mode with FM to do complex chirps as you wish.
2) use ArbBuilder, ArbExpress or your own script to create arbitrary waveforms. -
RE: TDS784A Letter of Volatility / Declassification guide.
Hi Daniel,
I've attached the letter of volatility with this post. -
RE: Oscilloscope DPO 7104C not booting
Hi Anandkumar,
I would suggest following the troubleshooting tree on page 4-22 of the service manual.
If that doesn't resolve the issue, your next option would be to send it in for repair.