  • RE: 2651A - Reading back listi values

    Fri, 17 Feb 2023 20:27:30 GMT – 2651A - Reading back listi values
    Hello,
    See page 7-34 in the reference manual attached below as this should give you a good idea on what you are wanting to accomplish.
    Model 2651A High Power System SourceMeter Instrument Reference Manual | Tektronix
    Also, I saw you set your source range to 10 but the last value in your lsit is 19. If the lsit does not autorange for you, this may cause some issues as you're attempting to source a value well outside of your set range of 10.
  • RE: DAQ6510 Display when using rear terminals under GPIB control

    Fri, 17 Feb 2023 15:53:41 GMT – DAQ6510 Display when using rear terminals under GPIB control
    Hello Dave,
    Are you operating via front panel or remotely and sending commands? If remote, which command set are you using (TSP or SCPI)?
    If using remote communication, you may need to add some display commands to specify what you want displayed on the front panel during data acquisition.
  • RE: L91 Battery Model for 2281S

    Wed, 15 Feb 2023 17:27:14 GMT – L91 Battery Model for 2281S
    Hello,
    Apologies, but we do not have that model generated. If you create a case with our technical support team we can discuss some options, if you are interested.
    Here is the link to create a case with us: Contact Tektronix Technical Support | Tektronix
  • RE: Possible methods for reducing data points for long term measurements?

    Tue, 14 Feb 2023 18:50:36 GMT – Possible methods for reducing data points for long term measurements?
    Hello,
    Can you create a case with our Technical Support Center? One of our Application Engineers will work with you one on one to help you find a solution. When you create the case, please add the link to this forum post for a reference. 
    Thanks!
    Contact Tektronix Technical Support | Tektronix
  • RE: Part number of mating 78-pin D-sub for 3723

    Fri, 10 Feb 2023 14:27:25 GMT – Part number of mating 78-pin D-sub for 3723
    Hello,
    We have an available connector you are looking for. Apologies for late response! See attached PNG.
    3720-MTC-378 PIN FEMALE TO MALE D-SUB CABLE ASSY, 3M
  • RE: KEITHLEY 2612B Source Meter SCPI Command Setup

    Wed, 01 Feb 2023 20:02:38 GMT – KEITHLEY 2612B Source Meter SCPI Command Setup
    Hello Eniola,
    The 26XX line of SMU's do not take SCPI commands. We utilize a command set known as TSP (test script processor) which operates in a similar way. I'll link the reference manual below, see page 2-51 and on for sweep functions and there are examples provided as you read along. Specifically, see section 9 for all TSP commands.
    Additionally, I would recommend downloading TSB (Test Script Builder) which will allow you to program and run commands remotely. In the navigator pane on the left hand side, you can find examples for a linear IV sweep example which would help you to start building and customizing your setup once you've read the sections I recommended above.

    Here is the reference manual: Series 2600B System SourceMeter Instrument Reference Manual | Tektronix

    Here is link to download TSB: Keithley Test Script Builder | Tektronix
  • RE: Zero check and Zero correct of 6514

    Tue, 31 Jan 2023 16:15:39 GMT – Zero check and Zero correct of 6514
    Hello,
    Zero check is a feature that notes the internal voltage offset within the unit. Zero correct allows you to algebraically subtract the internal voltage offset value from your measurement thereby correcting the value of zero within the instrument. Very similar to a REL function, however REL can be used to subtract out ANY level of signal or reading. Zero check/correct is specifically for correcting the value of zero by subtracting internal voltage offsets that would otherwise result in the instrument reading slightly higher than "true" zero.
    Zero Check and Zero correct should both be done prior to making a measurement. Zero check willl be disabled to perform a reading.
    See the manual I'm attaching below for more info. Ref pages 2-13 and on.
    Model 6514 System Electrometer Instruction Manual | Tektronix
  • RE: 2110 DMM Update problem

    Tue, 31 Jan 2023 15:56:49 GMT – 2110 DMM Update problem
    Kenneth,
    Did the error first show after trying to update firmware or was this before?
    What version were you on and which version were you upgrading to?
    We have seen some issues when updating from older firmware to newer even though there is no hardware mismatching. I'm afraid this unit may need to be sent in to service for our team to look in to if this error does not go away. I'll attach the service link below just in case:

    Use this link if located in the US: Service Quote | Tektronix
    Use this link if located OUTSIDE of the US:Request Instrument Repair/Calibration for Tektronix & Keithley Products | Tektronix
  • RE: KTH-2651A Minimum Operating Voltage as Load

    Fri, 20 Jan 2023 20:10:50 GMT – KTH-2651A Minimum Operating Voltage as Load
    Hello,
    1. Minimum operating voltage in 4th Quadrant as sink mode(or load mode)
    We have same voltage accuracy in source and sink as the attached table(see png). We have 5 voltage ranges 100mV to 40V. In the minimum range 100mV, we have 5uV programing resolution but +/-(0.02% of reading + 500uV offset) for uncertainty which means that we can set or program 5uV but actual value sources with +/- 500uV uncertainty. This is voltage source mode case (CV mode).
    The other case is that we can control output voltage level by compliance voltage in the current source mode. In this case, we have 10mV minimum in compliance voltage programing. I would not recommend the second case for accurate voltage sourcing applications.

     
    1. Pulse Condition
    Min/Max pulse width/rise time are very wide ranges depending on DUT impedance, source range and measure range. In the voltage source mode, we have usually about 300us ~ 1ms for pulse width. Please check out the datasheet for every individual case.
     
    1. Output Off mode
    If a DUT is an energy storage device like battery or solar cell, the output off mode must be high-Z mode. The default state of the instrument 2651A is 0V like ground. As soon as lead cables touch the DUT, the instrument should start to discharge the DUT. High-Z mode of output off mode prevents discharging issue in the DUT connection.  
     
  • RE: Part number of mating 78-pin D-sub for 3723

    Wed, 18 Jan 2023 19:35:50 GMT – Part number of mating 78-pin D-sub for 3723
    Hello,
    See attached link below and see bottom of page 2 for all parts and accessories for the 3723!
    Model 3723 Dual 1x30 High Speed Multiplexer Specifications | Tektronix
    Hope this helps
  • ‹