Hi
I have a problem with one 2602B, and it is wrong current measurement when used as voltage source of 5mA and complice set to 10mA (DUT current consumption is 1mA).
For 5V and compliance set to 10mA, SMU measures around 7uA.
For 5V and compliance set to 5mA, SMU measures the correct values of 1mA.
SMU returns no error, current is not limited, no blinking parameter
2w measurement.
SMU is connected directly to DUT Vcc without any external polarization circuit.
DUT entrance has a decoupling capacitor of 100nF between the two pins.
1mA should be measured correctly with both compliance configuration but it does not do it, could you explain why is this happening?
Tanks for quick reply.