web
You’re offline. This is a read only version of the page.
close
  • RE: 2600-PCT Curve tracer remediation and probe station integration?

    Hi Daniel,<br> <br> Interesting combination of products you have.&#160; Too bad the gate SMU is 2601B and not 2611B.<br> I'm conferring with factory about the 40V models and if the ~200V OVP is sufficient.<br> <br> <br> For the handling of the LO and Sense LO from the 2657A and the 2601B, we do have a &quot;ground unit&quot; accessory item:&#160; 2657A-LIM-3.<br> A single triax cable from the SLO/LO connector on each SMU is brought to the LIM-3.<br> The LIM-3 then provides output connectors for the combined LO and separately for the combined SLO.<br> the LIM-3 also provides interlock hub.<br> <br> <a href="https://www.tek.com/en/default-accessory-series-manual-29">Info on the 2657A-LIM-3</a><br> <br> Let me come back to you concerning the 40V model.<br> <br> Do you have any of the 2600-TRIAX accessory items to convert the 8-pin Phoenix connector to three triax for the gate SMU?<br> &#160;
  • RE: Entry Point Error in GPIB diagnoistic

    When you run into the trouble, find the gpib related DLL and right click to see properties; &#160;namely the author/publisher.<br> <br> You can have LabVIEW and other NI software (VISA) installed. &#160;Just don’t let the NI support for GPIB cards install, else it will overwrite the DLLs that need to be for your KPCI488-488A gpib interface card.<br> One strategy would be to install&#160;KPCI488-488A driver after other NI related software is installed.<br> &#160;
  • RE: DMM6500 streaming digitized data to computer via python + TSP code

    The appendix B code is configuring the instrument to use binary data format (format.data = format.REAL32).<br> Compared to ASCII, this allows fewer characters on the bus for better throughput.<br> <br> To interpret the data, checkout some IEEE754 compliant techniques.<br> I think numpy and fromfile might be useful.
  • RE: Keithley SMU 2450 Sample rate of one reading @ NPLC=0.01

    Model 2450 uses integrating A/D converter type.<br> It is custom implementation, but is reasonable to think of it as dual slope integrating A/D of about 22 bits at 1 NPLC.<br> The NPLC setting is affecting the duration of the integration or how long you let your signal “charge up” the cap.<br> Regardless of NPLC value, a reported measurement is based on sample size of one.<br> <br> In contrast, some of our products have both high accuracy integrating A/D and fast digitizing A/D.<br> The digitizer always runs at 1MHz; &#160;requested sampling rates slower than 1MHz can make use of auto aperture based averaging from the 1MHz data. &#160;SMU models such as 2461, 2651A have digitizers in addition to integrator.
  • RE: Cycle life of hot vs cold switching

    That sort of performance info would be specified by the OEM of the relay that is mounted into the S46 chassis.

    Assuming Radiall relay types, their document here may be of help:
    Radiall RF Relay Info

    On pg 14 of that document they give their recommendation about hot switching (don't).
    The estimated amount of degradation in the relay life/number of operations is a function of the average power of the signal.

    On pg. 9 of that document they give the Average Power Rating vs. Operating Frequency.
  • RE: Entry Point Error in GPIB diagnoistic

    Check your gpib-32.dll properties.  Is it from NI or Keithley or possibly CEC?

    My guess is that something installed the NI driver for GPIB.  And it over-wrote the required version of that DLL needed by the KPCI-488A card.
    Can you reinstall the driver for KPCI-488A?
  • RE: DAQ6510 Display when using rear terminals under GPIB control

    Did you close the channel?
    For the reading that is placed in the buffer, what channel tag does it have?

    My .NET code from prior post does cause the front panel to update the watch channel measured value.

    On the instrument, in the event log settings, you can enable command logging.  This will let you see exactly what the IVI driver commands are sending to the DAQ.

    If still not working, check your firmware level.  I'm at 1.7.12b

     
  • RE: Keithley 2614B : Unable to detect my SMU in Test script builder

    Check that you have the most recent version of Test Script Builder (J05).
    Also check the firmware on your 2612B (3.4.2 is most recent).

    Test Script Builder is making use of NI-VISA for communication with the instrument.
    From error screen, it is not clear what the issue is.

    If you have NI-MAX installed, try verifying basic VISA IO from that.

    If you still need help, tell us the VISA resource name that you use for the 2612B.
     
  • RE: 2461 SourceMeter pulse train clipping at 36 V short of 105 V spec

    What is the state of interlock?
    If interlock not satisfied, the max voltage is clipped/only a portion of 100V range can be used.
  • RE: DMM7510 ignores external trigger event

    Closing the loop on this one.

    Suggest updating firmware to latest (1.7.12b) ;  you are at 1.7.7b.

    Even so, when using the integrating A/D, we are able to duplicate the issue with DMM7510 but not with DMM6500.
    A bug has been logged for DMM7510 behavior.

    Workarounds if using DMM7510:

    First:  your trigger model operation is essentially sampling until told to stop;  You could get to essentially same outcome by making use of abort() command to stop an infinite count acquisition into a recycled finite sized buffer.

    Second:  your trigger model operation will work and *TRG can be detected/used to stop it if the digitizing A/D is used instead of the integrating A/D.

    Here is a code sample:
    my_instr.write("reset()")
    my_instr.write("errorqueue.clear()")
    
    #comment out the use of integrating A/D
    #my_instr.write("dmm.measure.func = dmm.FUNC_DC_VOLTAGE")
    #my_instr.write("dmm.measure.range = 0.1")
    
    ​​​​​​​#use the digitizing A/D
    my_instr.write("dmm.digitize.func = dmm.FUNC_DIGITIZE_VOLTAGE")  
    my_instr.write("dmm.digitize.range = 0.1")
    my_instr.write("dmm.digitize.samplerate = 10e3")
    my_instr.write("dmm.digitize.aperture = dmm.APERTURE_AUTO")
    
    my_instr.write("measure_buffer = buffer.make(12)")
    my_instr.write("measure_buffer.fillmode = buffer.FILL_CONTINUOUS")
    
    my_instr.write("trigger.model.setblock(1, trigger.BLOCK_BUFFER_CLEAR, measure_buffer)")
    my_instr.write("trigger.model.setblock(2, trigger.BLOCK_MEASURE_DIGITIZE, measure_buffer, trigger.COUNT_INFINITE)")
    my_instr.write("trigger.model.setblock(3, trigger.BLOCK_WAIT, trigger.EVENT_COMMAND, \
                                                  trigger.CLEAR_ENTER, \
                                                  trigger.WAIT_OR, \
                                                  trigger.EVENT_DISPLAY)")
    my_instr.write("trigger.model.setblock(4, trigger.BLOCK_MEASURE_DIGITIZE, measure_buffer, trigger.COUNT_STOP)")