Hello,<br>
<br>
The HV-CS-1613 is just a feedthrough or barrel type connector utilizing our HV Triax connector type on both ends, so do not think it will be what you want either.<br>
<br>
For the double pulse test, I suspect you are using the SMU just to charge up the capacitor bank, right?<br>
I imagine that measuring very low currents with the 2657A is not important to this application.<br>
Therefore, you could make use of two cables of model SHV-CA-553-x (x is the length in meters).<br>
One end of this cable has the proprietary HV Triax connector that will connect with the HI or LO at rear of 2657A.<br>
The other end is an SHV Plug connector.<br>
Your PCB or custom test fixture could then make use of commonly available SHV Jack connectors.<br>
Google 'SHV Panel Mount' to get an idea of what the open market offers for SHV.<br>
<br>
<br>
You mention having an 8010 test fixture and the HV-CA-554 cables already to connect the 2657A to the 8010.<br>
If you want to continue to use those cables, but you need a means to connect to a custom test fixture or PCB board rather than to the 8010, consider using the HV-CA-571-3 feedthrough with unterminated triax.<br>
Alternately, there is a PCB mount HV-CS-1589 that might also work for your situation.
-
Just confirming you have the older 3706 and not the 3706<strong>A?</strong><br>
<br>
 Assuming 3706, page 7-2 of User Manual has the upgrade instructions.<br>
(https://www.tek.com/en/switching-and-data-acquisition-systems/3700a-systems-switch-multimeter-manual-3)<br>
<br>
With the thumb drive inserted, you need to use the menu to navigate to the UPGRADE screen.  Then the content from the thumb drive will be applied.<br>
The two files that you placed will cause any "autoexec" application of the firmware update.
-
Can you clarify which forum post gives you this error?<br>
The posts you have pertaining to Clarius install seem fine to me;   not errors to display.
-
You may want to revisit the idea of using 2W for the high ohms channel.<br>
When a measure function is associated with a channel, then when closing that channel, all the required channel pairing (in case of 4W) and backplane relays are managed for you.<br>
So a scan that uses both 2W and 4W will mange the backplanes appropriately for the closed channel.<br>
<br>
If using a scan with mixed measure function type, I'm able to reliably measure 50MΩ (2Wire) and 0.5Ω (4Wire) on two different channels.  See attached.<br>
<br>
If not performing a scan and attempting use of 4W with the 50MΩ, I find extra settling time is needed.<br>
I'm not able to find a command in the IVI driver.  So using the WriteString() to send:  ":ROUT:DEL 0.25, (@106)"  to obtain an extra 250msec on channel 106.<br>
<br>
See attached for info on both defining a scan or the single channel read approach.
-
If you want to connect the two meters together without soldering a cable, need three items:<br>
triax cable to 6430 preamp (7078-TRX-3)<br>
Insert 237-BNC-TRX into end of 7078-TRX-3 cable<br>
Put 7078-TRX-BNC triax to BNC adapter onto 6517B;  attach the cable.<br>
<br>
If you want to make your own, see attached image.
-
How long are your cables from the 7707 and the devices?<br>
<br>
Also, for the high ohm channel, can you instead use 2W ohms mode?<br>
I suspect part of the settling time issue will be for charging up the remote sense lines.<br>
For 10M-Ohm and higher DUT value, the use of 4W is of negligible benefit and in this case might be giving you a longer settling time problem.
-
For topic #2, check the state of your interlock.  I suspect that is not engaged and it is normal for the instrument to clamp the output below ~42V.  In your case, you are hitting the voltage clamp level and so the current forcing is not the primary action of the SMU any longer;  the actual current will be the clamped V / DUT R.<br>
<br>
For topic #1, error 5111 is because the measure time will exceed the requested pulse width.<br>
Try setting the NPLC to minimum of 0.01 which for 60Hz power equates to 166.7usec.<br>
Make sure you turn off the source readback;  that defaults to ON which means two measurements occur (one for V and one for I) which at 0.01 NPLC puts you past the 250usec pulse width.  Turn it off so that if forcing I, it only measures V and does not also read back the source level.<br>
Source or Measure Delays could also be playing a role;  set them to zero.<br>
<br>
Last:  make sure your firmware level is up to date.  1.7.12b is current release.
-
Hi Daniel,<br>
<br>
Interesting combination of products you have.  Too bad the gate SMU is 2601B and not 2611B.<br>
I'm conferring with factory about the 40V models and if the ~200V OVP is sufficient.<br>
<br>
<br>
For the handling of the LO and Sense LO from the 2657A and the 2601B, we do have a "ground unit" accessory item:  2657A-LIM-3.<br>
A single triax cable from the SLO/LO connector on each SMU is brought to the LIM-3.<br>
The LIM-3 then provides output connectors for the combined LO and separately for the combined SLO.<br>
the LIM-3 also provides interlock hub.<br>
<br>
<a href="https://www.tek.com/en/default-accessory-series-manual-29">Info on the 2657A-LIM-3</a><br>
<br>
Let me come back to you concerning the 40V model.<br>
<br>
Do you have any of the 2600-TRIAX accessory items to convert the 8-pin Phoenix connector to three triax for the gate SMU?<br>
 
-
When you run into the trouble, find the gpib related DLL and right click to see properties;  namely the author/publisher.<br>
<br>
You can have LabVIEW and other NI software (VISA) installed.  Just don’t let the NI support for GPIB cards install, else it will overwrite the DLLs that need to be for your KPCI488-488A gpib interface card.<br>
One strategy would be to install KPCI488-488A driver after other NI related software is installed.<br>
 
-
The appendix B code is configuring the instrument to use binary data format (format.data = format.REAL32).<br>
Compared to ASCII, this allows fewer characters on the bus for better throughput.<br>
<br>
To interpret the data, checkout some IEEE754 compliant techniques.<br>
I think numpy and fromfile might be useful.