• RE: Using 2657A for Double Pulse Test

    Hello,<br> <br> The HV-CS-1613 is just a feedthrough or barrel type connector utilizing our HV Triax connector type on both ends, so do not think it will be what you want either.<br> <br> For the double pulse test, I suspect you are using the SMU just to charge up the capacitor bank, right?<br> I imagine that measuring very low currents with the 2657A is not important to this application.<br> Therefore, you could make use of two cables of model SHV-CA-553-x (x is the length in meters).<br> One end of this cable has the proprietary HV Triax connector that will connect with the HI or LO at rear of 2657A.<br> The other end is an SHV Plug connector.<br> Your PCB or custom test fixture could then make use of commonly available SHV Jack connectors.<br> Google 'SHV Panel Mount' to get an idea of what the open market offers for SHV.<br> <br> <br> You mention having an 8010 test fixture and the HV-CA-554 cables already to connect the 2657A to the 8010.<br> If you want to continue to use those cables, but you need a means to connect to a custom test fixture or PCB board rather than to the 8010, consider using the HV-CA-571-3 feedthrough with unterminated triax.<br> Alternately, there is a PCB mount HV-CS-1589 that might also work for your situation.
  • RE: 3706 upgrade procedure

    Just confirming you have the older 3706 and not the 3706<strong>A?</strong><br> <br> &#160;Assuming 3706, page 7-2 of User Manual has the upgrade instructions.<br> (https://www.tek.com/en/switching-and-data-acquisition-systems/3700a-systems-switch-multimeter-manual-3)<br> <br> With the thumb drive inserted, you need to use the menu to navigate to the UPGRADE screen.&#160; Then the content from the thumb drive will be applied.<br> The two files that you placed will cause any &quot;autoexec&quot; application of the firmware update.
  • RE: I can't see...

    Can you clarify which forum post gives you this error?<br> The posts you have pertaining to Clarius install seem fine to me;&#160; &#160;not errors to display.
  • RE: Switching between Resistance Reading Channels on DAQ6510

    You may want to revisit the idea of using 2W for the high ohms channel.<br> When a measure function is associated with a channel, then when closing that channel, all the required channel pairing (in case of 4W) and backplane relays are managed for you.<br> So a scan that uses both 2W and 4W will mange the backplanes appropriately for the closed channel.<br> <br> If using a scan with mixed measure function type, I'm able to reliably measure 50MΩ (2Wire) and 0.5Ω (4Wire) on two different channels.&#160; See attached.<br> <br> If not performing a scan and attempting use of 4W with the 50MΩ, I find extra settling time is needed.<br> I'm not able to find a command in the IVI driver.&#160; So using the WriteString() to send:&#160; &quot;:ROUT:DEL 0.25, (@106)&quot;&#160; to obtain an extra 250msec on channel 106.<br> <br> See attached for info on both defining a scan or the single channel read approach.
  • RE: question about Keithley 6430 connect to 6517B

    If you want to connect the two meters together without soldering a cable, need three items:<br> triax cable to 6430 preamp (7078-TRX-3)<br> Insert 237-BNC-TRX into end of 7078-TRX-3 cable<br> Put 7078-TRX-BNC triax to BNC adapter onto 6517B;&#160; attach the cable.<br> <br> If you want to make your own, see attached image.
  • RE: Switching between Resistance Reading Channels on DAQ6510

    How long are your cables from the 7707 and the devices?<br> <br> Also, for the high ohm channel, can you instead use 2W ohms mode?<br> I suspect part of the settling time issue will be for charging up the remote sense lines.<br> For 10M-Ohm and higher DUT value, the use of 4W is of negligible benefit and in this case might be giving you a longer settling time problem.
  • RE: Error 5111 on Keithley 2461

    For topic #2, check the state of your interlock.&#160; I suspect that is not engaged and it is normal for the instrument to clamp the output below ~42V.&#160; In your case, you are hitting the voltage clamp level and so the current forcing is not the primary action of the SMU any longer;&#160; the actual current will be the clamped V / DUT R.<br> <br> For topic #1, error 5111 is because the measure time will exceed the requested pulse width.<br> Try setting the NPLC to minimum of 0.01 which for 60Hz power equates to 166.7usec.<br> Make sure you turn off the source readback;&#160; that defaults to ON which means two measurements occur (one for V and one for I) which at 0.01 NPLC puts you past the 250usec pulse width.&#160; Turn it off so that if forcing I, it only measures V and does not also read back the source level.<br> Source or Measure Delays could also be playing a role;&#160; set them to zero.<br> <br> Last:&#160; make sure your firmware level is up to date.&#160; 1.7.12b is current release.
  • RE: 2600-PCT Curve tracer remediation and probe station integration?

    Hi Daniel,<br> <br> Interesting combination of products you have.&#160; Too bad the gate SMU is 2601B and not 2611B.<br> I'm conferring with factory about the 40V models and if the ~200V OVP is sufficient.<br> <br> <br> For the handling of the LO and Sense LO from the 2657A and the 2601B, we do have a &quot;ground unit&quot; accessory item:&#160; 2657A-LIM-3.<br> A single triax cable from the SLO/LO connector on each SMU is brought to the LIM-3.<br> The LIM-3 then provides output connectors for the combined LO and separately for the combined SLO.<br> the LIM-3 also provides interlock hub.<br> <br> <a href="https://www.tek.com/en/default-accessory-series-manual-29">Info on the 2657A-LIM-3</a><br> <br> Let me come back to you concerning the 40V model.<br> <br> Do you have any of the 2600-TRIAX accessory items to convert the 8-pin Phoenix connector to three triax for the gate SMU?<br> &#160;
  • RE: Entry Point Error in GPIB diagnoistic

    When you run into the trouble, find the gpib related DLL and right click to see properties; &#160;namely the author/publisher.<br> <br> You can have LabVIEW and other NI software (VISA) installed. &#160;Just don’t let the NI support for GPIB cards install, else it will overwrite the DLLs that need to be for your KPCI488-488A gpib interface card.<br> One strategy would be to install&#160;KPCI488-488A driver after other NI related software is installed.<br> &#160;
  • RE: DMM6500 streaming digitized data to computer via python + TSP code

    The appendix B code is configuring the instrument to use binary data format (format.data = format.REAL32).<br> Compared to ASCII, this allows fewer characters on the bus for better throughput.<br> <br> To interpret the data, checkout some IEEE754 compliant techniques.<br> I think numpy and fromfile might be useful.