Try setting the source level and source limits before turning the output on.
Keep in mind, when using power limit, the current limit in use will be the lower of the limit.i value or the computed value from source v and requested power limit (watts).
If troubles persist, can you tell us more about the DUT error and the only if device disconnected observations.
Is device accepting power from SMU or is SMU sinking power from DUT?
-
KickStart can only scan the channels.
If ch1 has 5v and gnd1 and ch2 has 5v and 12v and so on.
Otherwise, you’d need some customized approach to try to measure between any two pins.
That is normally done with a matrix rather than multiplexing switch.
-
You need to assign the range command also to the channel. Similar to what you did for the measure function.
-
The data sheet indicates:
Maximum input: 250 Vpeak DC to 60 Hz sine wave; 10 s per minute at maximum input on mA ranges (one-sixth
duty cycle)
Maximum common mode voltage (DC to 60 Hz sine wave): Electrometer, 500 Vpeak; V-source, 750 Vpeak
You may want to mount some protective diodes. See attached from our Low Level Measurements Handbook.
Or download a copy:
Low Level Measurements Handbook
-
Can you let us know which version of KickStart you use?
-
The SourceMeter does not have slew rate or rise time control.
Instead, it implements a smooth first order response as it transitions to a new source level.
This allows best chance for no overshoots or ringing as it reaches new steady state source levels.
-
Hi,
Are you working with only two terminal devices? In your original post it seems that might be the case.
Do you need to use more than one test frequency?
As you have discovered, when using bias tee, the mode needs to be managed during collection of the compensation values. For nF and smaller, open compensation is most important.
You can accomplish that with just subtracting the open ckt measured value from your DUT measurement (at same test frequency).
The KXCI commands for collection and application of the compensation to measurements is implementing the same functions that Clarius gives for collection and use of compensation. However, these are not expecting the bias tees.
Most of our 3KV HVCV installations are for three terminal devices (Ciss, Coss, Crss). For this, we need different compensation values for each of the three measurements (due to cabling differences). This influenced the ACS software implementation approach.
I have concerns about stacking up bias tees to address the issue.
-
When using the scan related commands, only one channel relay will be closed at a time.
If two or more channels are closed, the measured value will the combined ohms.
If the multiple closed channels is really what you need, you can close channels and separately command the dmm to measure. You will need to also manage the backplane relays to connect the dmm to the switch hi and lo.
-
Are there any differences in the length of cables for your 20-channel setup vs. this one with only 5 channels?
If longer delay is allowing a good outcome, it implies a difference in the time constant of your DUT.
Does it stop reporting the overflow if you turn off the open lead detection? The reported overflow could come a saturated measure range or from a detected open lead. Since more delay fixes the overflow, this implies it is coming from the measurement from after the open lead detection has occurred.
To look into it, you might put a scope on the Card Sense HI and LO terminals. When doing 4-wire ohms, the current is forced on the HI and LO and the resulting I*R is measured on the Sense HI and LO.
Is that spiking and requiring more settling time vs hte 20-channels setup?
-
For control of 3706A from .NET (C#), I'd suggest:
approach 1: use a VISA layer and directly send the instrument commands (3706A Reference Manual)
approach 2: use IVI driver. This will leverage NI VISA runtime to communicate with the 3706A; driver commands are translated to the instrument commands for you.
Both NI VISA and the IVI driver distribute some sample code in C#.