web
Вы не подключены к сети. Эта версия страницы доступна только для чтения.
close
  • RE: [Ivi.Visa] Time Out Error when commnunicate with 4200A through GPIB

    When using LAN based connection to KXCI, it is normal for every config command you send to respond back with ACK\0.
    Pg 2-11 in the KXCI Remote Control Programming manual shows ethernet based write and read examples.

    For example, sending ME\0 command, the 4200A will respond back with ACK\0

    In your code, if using LAN to KXCI, plan to do a write + read for every operation to pull the ACK\0 off the bus.
  • RE: Python for 4200-SCS

    On both the older 4200-SCS that you have and our 4200A-SCS, there is the KXCI interface.
    This is the Keithley eXternal Command Interface or KXCI.
    Close KITE or Clarius and invoke KXCI.
    But first visit KCON and config the LAN or GPIB interface.

    Attached is an application note on using Python with 4200A-SCS.

     
  • RE: Keithley 2400 - Voltage Offset

    An offset of 700mV is quite large for any of the ranges.
    I'd say it is unlikely to be a calibration issue.  It most likely requires some service.

    Humor us and double check the setup.  From power on default behavior, if you just turn the output on (blue light) it would be configured to source 0V with current limit of 105uA.
    Do any digits on front panel flash rapidly?
    And if you connected a voltmeter to the HI and LO, you see 0.7 volts instead of 0V (don't use the resistor)?
    Is it a constant amount of offset....if you asked the 2400 to source 1V, would you get 1.7V?
  • RE: Trouble with Ethernet-connected Keithley lab bench setup

    For the 2380 load, the only application of KickStart that recognizes the load is the Battery app.  That is an extra license (not included in -SUITE).
    It can control the load for generating a battery model.

    The 2260B may only be recognized by KickStart when using the USB-CDC interface not the LAN.
  • RE: K2400 - Voltage fluctuations

    You should probably talk this over with your local applications engineer.

    What is the maximum short circuit current from your device?

    I am guessing the SMU is used in force voltage and limit current mode with your solar PN junction.
    You might be encountering current compliance or range compliance.  Anything on the front panel of 2400 go to flashing on/off when you open the shutter?
  • RE: [4200A -CVU] Prolem with CV measurement at 1Khz

    Can you explain the 3-sigma value?  How compute it?

    In your config, what sort of speed settings?  Use Quiet if looking for small standard deviations.

    It looks like your DUT value is a few single digit pF.
    At 1KHz, it will be very big Z and very small current relative to using higher test freq.
    In our datasheet for CVU, we do not publish value for pF at 1KHz.

     
  • RE: I want to attach my DPO4000B oscilloscope with SCS 4200. How can i do that?

    Sure, it is possible.

    For older 4200, they can connect to external instruments on gpib or LAN (raw socket).
    The newer 4200A-SCS adds usb and VISA for connection to external instruments.

    The scope you mention has both LAN and USB.

    For the software environment to send commands to scope and read data back, a library would be written using KULT tool.  Depending on your overall objectives, the code you write with KULT can also interact with the SMU, CVU or PMU hardware in your 4200/4200A.  The compiled library is run from KITE or Clarius UTM (user test module).  KXCI can also call user libraries.

    Take a look at example KULT libraries on your system.  Ones for temperature control or for external LCR meter can give you an idea of interaction with external instruments.
     
  • RE: 2602B channels in parallel for increased current

    Yes, that can be done.

    In general, follow the Guidelines for combining SMU outputs in the Reference manual (section 2).

    For two current sources in particular, make sure to have the voltage compliance limits a little bit different from each other.

    To have same timing behavior from each SMU, use same ranges and other settings on each.

    Also, sometimes the order of starting their trigger models matters.  For example, if SMUB is consuming an output event from SMUA to get things started, make sure to call the trigger initiate for SMUB first so that it can be fully armed and awaiting the events.
  • RE: Discontinuity on IV graph during voltage sweep on diode

    Are you using a fixed source range?  If the source range is also using auto, then the discontinuity could be for the hand-off of 200mV to 2V range as you sweep to 250mV.  A source range can go to 101% of its full-scale value.

    For the data in the buffer, look at the measurerange and sourcerange attributes.  And perhaps post a CSV with your I, V, timestamp and the range information.
    Try setting the smu sourcerange to a fixed value that can handle all the desired source levels (2V in your case).

    When was last time your instrument was calibrated?
     
  • RE: Control Voltage for C-V Measurements with 2600-RBT-200 Bias Tee (CVU-3K-KIT)

    The CVU can source the DC bias at both CVUHI and CVULO.
    To put the bias tees into CV mode, source -10 at each terminal from CVU.
    The blocking cap of the bias tee will prevent this -10V from reaching DUT;  instead the bias from the SMU passes thru.

    You are writing custom code to control it all?
    Our ACS or ACS-Basic software has HVCV libraries for this configuration.