RE: [4200A-SCS] KXCI recieve wrong command
In KCON settings, your EOI is off. Try turning that on in KCON for the gpib communication. -
RE: Remote Restart SMU
Thanks for clarifying.
see if this post helps. Same DST process for SMU or DAQ:
https://my.tek.com/en/tektalk/switching-data-acq/8644f450-3c70-ee11-a81c-000d3a32997c
-
RE: How can I set "pulse mode" in keithley 2602B ?
Needs an update:
The code posted above will not take you into pulse region 4.
These high watt pulse regions are in our Extended Operating Area and have firmware imposed max PW and Duty Cycle limits.
To access the EOA regions, supply a higher voltage value for the trigger.source.limitv which will be applied only during the high watt pulse.
When not sourcing the pulse level, the lower value set for the smuX.source.limitv applies. The combo of smuX.source.leveli and smuX.source.limitv needs to be within the DC operating area of the SMU.
These three lines in particular for accessing EOA region:
smuX.measure.rangev = EOA_limit
smuX.source.limitv = limitV
smuX.trigger.source.limitv = EOA_limit
I'll upload scope shot of 1.5A into 10Ω.
Revised code:function Run_test(show_data, PulsePeriod, PulseWidth, MeasDelay) -- when passthrough is true, the timer will "tic" immediately, -- which for this application will start as soon as the -- trigger model is ARMED. node[1].trigger.timer[1].delay = PulsePeriod node[1].trigger.timer[1].stimulus = node[1].smua.trigger.ARMED_EVENT_ID node[1].trigger.timer[1].count = 1 --PulseCount - 1 node[1].trigger.timer[1].passthrough = true node[1].trigger.timer[1].clear() node[1].trigger.timer[2].delay = PulseWidth node[1].trigger.timer[2].stimulus = node[1].trigger.timer[1].EVENT_ID node[1].trigger.timer[2].count = 1 node[1].trigger.timer[2].passthrough = false node[1].trigger.timer[2].clear() node[1].trigger.timer[3].delay = MeasDelay node[1].trigger.timer[3].stimulus = node[1].trigger.timer[1].EVENT_ID node[1].trigger.timer[3].count = 1 node[1].trigger.timer[3].passthrough = false node[1].trigger.timer[3].clear() -- digital output trigger for external equipment -- configure digital IO line 1 to output a active LO/falling edge -- pulse at start of each current pulse digio.trigger[1].clear() digio.trigger[1].mode = digio.TRIG_FALLING --digio.TRIG_RISINGM digio.trigger[1].pulsewidth = 100e-6 digio.trigger[1].stimulus = node[1].trigger.timer[1].EVENT_ID -- source stimulus too -- clear the reading buffers reset_buffers(node[1].smua) -- Turn the outputs on smua.source.output = smua.OUTPUT_ON -- after blue light state change, delay a little delay(0.1) -- Start the trigger model execution smua.trigger.initiate() -- start this one last as all the other smus follow this leader -- Wait until the sweep has completed waitcomplete() --[[ -- hot switch: -- to bleed off the trapped charge on PN junction capacitance and non-zero voltage -- briefly set the voltage compliance to small number restore_val = smua.source.limitv smua.source.limitv = 0.01 delay(0.001) smua.source.limitv = restore_val ]] -- before blue light state change, delay a little delay(0.1) -- Turn the output off smua.source.output = smua.OUTPUT_OFF -- IV data for the laser if show_data == true then -- Print the data back to the Console in tabular format print(" Voltage\tCurrent\tResistance") for x=1,smua.nvbuffer1.n do -- Voltage readings are in nvbuffer2. Current readings are in nvbuffer1. print(smua.nvbuffer2[x], smua.nvbuffer1[x], smua.nvbuffer2[x]/smua.nvbuffer1[x] ) end end end function reset_buffers(smu) -- Prepare the Reading Buffers smu.nvbuffer1.clear() smu.nvbuffer1.appendmode = 1 smu.nvbuffer1.collecttimestamps = 1 smu.nvbuffer2.clear() smu.nvbuffer2.appendmode = 1 smu.nvbuffer2.collecttimestamps = 1 end function config_smu_list_sweep(nodenum, smu, srcRange, srclist, limitV, EOA_limit, nplc, remoteSense) smu.source.func = smu.OUTPUT_DCAMPS if (remoteSense == true) then smu.sense = smu.SENSE_REMOTE else smu.sense = smu.SENSE_LOCAL end smu.source.autorangei = smu.AUTORANGE_OFF smu.source.rangei = srcRange smu.source.leveli = 0 smu.source.limitv = limitV -- Disabling Auto-Ranging and Auto-Zero ensures accurate and consistent timing smu.measure.autozero = smu.AUTOZERO_ONCE smu.measure.autorangev = smu.AUTORANGE_OFF smu.measure.rangev = EOA_limit smu.measure.nplc = nplc -- A timer will be used to set the measure delay and synchronize the measurement -- between the SMUs so set the built in delay to 0. smu.measure.delay = 0 -- Configure SMU Trigger Model for Sweep --smu.trigger.source.lineari(start, stop, numPoints) smu.trigger.source.listi(srclist) smu.trigger.source.limitv = EOA_limit smu.trigger.measure.action = smu.ENABLE smu.trigger.measure.iv(smu.nvbuffer1, smu.nvbuffer2) smu.trigger.endpulse.action = smu.SOURCE_IDLE smu.trigger.endsweep.action = smu.SOURCE_IDLE smu.trigger.count = table.getn(srclist) smu.trigger.arm.stimulus = 0 smu.trigger.source.stimulus = trigger.timer[1].EVENT_ID smu.trigger.measure.stimulus = trigger.timer[3].EVENT_ID smu.trigger.endpulse.stimulus = trigger.timer[2].EVENT_ID smu.trigger.source.action = smu.ENABLE end -- ************************************** MAIN PROGRAM BELOW ***************** reset() errorqueue.clear() nodesFound = tsplink.reset(1) print("Nodes found = " .. nodesFound) --smua.interlock.enable = smua.DISABLE -- use of pulse region 4 of 2602B: 10A at up to 20V -- max 1.8msec pulse with 1% duty cycle local PulseWidth = 1.0e-3 local PulsePeriod = PulseWidth / 0.01 local MeasDelay = PulseWidth * 0.8 local nplc = (PulseWidth - MeasDelay - 60e-6) * localnode.linefreq print("Computed NPLC value: "..tostring(nplc)) if tostring(nplc) < "0.001" then nplc = 0.001 -- 16.7usec, minimum value allowed MeasDelay = 1e-6 end -- if -- voltage ranges on 2602B: 100mV, 1V, 6V, 40V -- pick a limit value higher than expected I*DUT_R and -- puts you on appropriate range. local voltageLimit = 1 local duringPulse_voltageLimit = 20 -- ranges: decades of 10.....100mA, 1A, 3A, 10A local mySrcRange = 3 mysrclist = { 1.5} -- use this one for a single pulse --local numPulses = table.getn(mysrclist) --config_smu_list_sweep(nodenum, smu, srcRange, srclist, limitV, EOA_limit, nplc, remoteSense) config_smu_list_sweep(1, node[1].smua, mySrcRange, mysrclist, voltageLimit, duringPulse_voltageLimit, nplc, true) timer.reset() Run_test(true, PulsePeriod, PulseWidth, MeasDelay) -- pass true to print data back to console time = timer.measure.t() print("Time to run: "..time) print("****************************************") print("Current Source Range: "..node[1].smua.source.rangei)
-
RE: Driver to control Keithley 3706A with visual basic application using ethernet cable
Attached is a form file from a VB sample using IVI driver and 3706A. -
RE: Driver to control Keithley 3706A with visual basic application using ethernet cable
The IVI driver is probably the way to go.
The IVI driver "sits upon" NI VISA for the communication with instruments.
Instead of GPIB0:: type resource name, pass info for your LAN connected instrument.
Typically like this: TCPIP0::192.168.1.50::inst0::INSTR
Substitute your IP address in there. -
RE: How can I set "pulse mode" in keithley 2602B ?
See how this does for you.
In the attachment is scope shot of the 1.5Amp, 1msec pulse into 50mΩ resistor (~75mV response).
TSP code: function Run_test(show_data, PulsePeriod, PulseWidth, MeasDelay) -- when passthrough is true, the timer will "tic" immediately, -- which for this application will start as soon as the -- trigger model is ARMED. node[1].trigger.timer[1].delay = PulsePeriod node[1].trigger.timer[1].stimulus = node[1].smua.trigger.ARMED_EVENT_ID node[1].trigger.timer[1].count = 1 --PulseCount - 1 node[1].trigger.timer[1].passthrough = true node[1].trigger.timer[1].clear() node[1].trigger.timer[2].delay = PulseWidth node[1].trigger.timer[2].stimulus = node[1].trigger.timer[1].EVENT_ID node[1].trigger.timer[2].count = 1 node[1].trigger.timer[2].passthrough = false node[1].trigger.timer[2].clear() node[1].trigger.timer[3].delay = MeasDelay node[1].trigger.timer[3].stimulus = node[1].trigger.timer[1].EVENT_ID node[1].trigger.timer[3].count = 1 node[1].trigger.timer[3].passthrough = false node[1].trigger.timer[3].clear() -- digital output trigger for external equipment -- configure digital IO line 1 to output a active LO/falling edge -- pulse at start of each current pulse digio.trigger[1].clear() digio.trigger[1].mode = digio.TRIG_FALLING --digio.TRIG_RISINGM digio.trigger[1].pulsewidth = 100e-6 digio.trigger[1].stimulus = node[1].trigger.timer[1].EVENT_ID -- source stimulus too -- clear the reading buffers reset_buffers(node[1].smua) -- Turn the outputs on smua.source.output = smua.OUTPUT_ON -- after blue light state change, delay a little delay(0.1) -- Start the trigger model execution smua.trigger.initiate() -- start this one last as all the other smus follow this leader -- Wait until the sweep has completed waitcomplete() --[[ -- hot switch: -- to bleed off the trapped charge on PN junction capacitance and non-zero voltage -- briefly set the voltage compliance to small number restore_val = smua.source.limitv smua.source.limitv = 0.01 delay(0.001) smua.source.limitv = restore_val ]] -- before blue light state change, delay a little delay(0.1) -- Turn the output off smua.source.output = smua.OUTPUT_OFF -- IV data for the laser if show_data == true then -- Print the data back to the Console in tabular format print(" Voltage\tCurrent\tResistance") for x=1,smua.nvbuffer1.n do -- Voltage readings are in nvbuffer2. Current readings are in nvbuffer1. print(smua.nvbuffer2[x], smua.nvbuffer1[x], smua.nvbuffer2[x]/smua.nvbuffer1[x] ) end end end function reset_buffers(smu) -- Prepare the Reading Buffers smu.nvbuffer1.clear() smu.nvbuffer1.appendmode = 1 smu.nvbuffer1.collecttimestamps = 1 smu.nvbuffer2.clear() smu.nvbuffer2.appendmode = 1 smu.nvbuffer2.collecttimestamps = 1 end function config_smu_list_sweep(nodenum, smu, srcRange, srclist, limitV, nplc, remoteSense) smu.source.func = smu.OUTPUT_DCAMPS if (remoteSense == true) then smu.sense = smu.SENSE_REMOTE else smu.sense = smu.SENSE_LOCAL end smu.source.autorangei = smu.AUTORANGE_OFF smu.source.rangei = srcRange smu.source.leveli = 0 smu.source.limitv = limitV -- Disabling Auto-Ranging and Auto-Zero ensures accurate and consistent timing smu.measure.autozero = smu.AUTOZERO_ONCE smu.measure.autorangev = smu.AUTORANGE_OFF smu.measure.rangev = limitV smu.measure.nplc = nplc -- A timer will be used to set the measure delay and synchronize the measurement -- between the SMUs so set the built in delay to 0. smu.measure.delay = 0 -- Configure SMU Trigger Model for Sweep --smu.trigger.source.lineari(start, stop, numPoints) smu.trigger.source.listi(srclist) smu.trigger.source.limitv = limitV smu.trigger.measure.action = smu.ENABLE smu.trigger.measure.iv(smu.nvbuffer1, smu.nvbuffer2) smu.trigger.endpulse.action = smu.SOURCE_IDLE smu.trigger.endsweep.action = smu.SOURCE_IDLE smu.trigger.count = table.getn(srclist) smu.trigger.arm.stimulus = 0 smu.trigger.source.stimulus = trigger.timer[1].EVENT_ID smu.trigger.measure.stimulus = trigger.timer[3].EVENT_ID smu.trigger.endpulse.stimulus = trigger.timer[2].EVENT_ID smu.trigger.source.action = smu.ENABLE end -- ************************************** MAIN PROGRAM BELOW ***************** reset() errorqueue.clear() nodesFound = tsplink.reset(1) print("Nodes found = " .. nodesFound) --smua.interlock.enable = smua.DISABLE -- use of pulse region 4 of 2602B: 10A at up to 20V -- max 1.8msec pulse with 1% duty cycle local PulseWidth = 1.0e-3 local PulsePeriod = PulseWidth / 0.01 local MeasDelay = PulseWidth * 0.8 local nplc = (PulseWidth - MeasDelay - 60e-6) * localnode.linefreq print("Computed NPLC value: "..tostring(nplc)) if tostring(nplc) < "0.001" then nplc = 0.001 -- 16.7usec, minimum value allowed MeasDelay = 1e-6 end -- if -- voltage ranges on 2602B: 100mV, 1V, 6V, 40V -- pick a limit value higher than expected I*DUT_R and -- puts you on appropriate range. local voltageLimit = 1 -- ranges: decades of 10.....100mA, 1A, 3A, 10A local mySrcRange = 3 mysrclist = { 1.5} -- use this one for a single pulse --local numPulses = table.getn(mysrclist) --config_smu_list_sweep(nodenum, smu, srcRange, srclist, limitV, nplc, remoteSense) config_smu_list_sweep(1, node[1].smua, mySrcRange, mysrclist, voltageLimit, nplc, true) timer.reset() Run_test(true, PulsePeriod, PulseWidth, MeasDelay) -- pass true to print data back to console time = timer.measure.t() print("Time to run: "..time) print("****************************************") print("Current Source Range: "..node[1].smua.source.rangei)
-
RE: Triggering SMUs via tsplink
See if this helps:--[[ Purpose: show how an event on one 2600B can be "echoed" to second 2600B Setup: need two 2600B that have TSP-LINK connectors. Assign node 1 and 2 and use the cross-over cable to link them. On front panel, use the reset under TSPLINK to verify the link. Method: Use a timer as stimulus for digital IO active lo pulse. Cause 10 pulses from DIO line 1 on rear DB25 connector. Echo the timer event_id on tsp-link for digital IO on second node to also pulse. ]]-- function ConfigTSPLinkTriggers(nodenum) node[nodenum].tsplink.trigger[1].clear() node[nodenum].tsplink.trigger[1].mode = tsplink.TRIG_FALLING node[nodenum].tsplink.trigger[2].clear() node[nodenum].tsplink.trigger[2].mode = tsplink.TRIG_FALLING node[nodenum].tsplink.trigger[3].clear() node[nodenum].tsplink.trigger[3].mode = tsplink.TRIG_FALLING end -- ********************* -- -- Main Program -- -- ********************* reset() errorqueue.clear() if tsplink.state == "offline" then nodesFound = tsplink.reset() if nodesFound ~= 2 then print(string.format("Error: Found %d Nodes. Expecting 2.", nodesFound)) exit() end end -- for each tsplink node ConfigTSPLinkTriggers(1) ConfigTSPLinkTriggers(2) -- configure a timer on node 1 to issue 10 events -- upon receipt of one tsplink event_id nodenum = 1 node[nodenum].trigger.timer[1].delay = 1e-3 node[nodenum].trigger.timer[1].stimulus = node[nodenum].tsplink.trigger[1].EVENT_ID node[nodenum].trigger.timer[1].count = 10 node[nodenum].trigger.timer[1].passthrough = false node[nodenum].trigger.timer[1].clear() -- echo the timer event on tsplink two node[nodenum].tsplink.trigger[2].stimulus = node[nodenum].trigger.timer[1].EVENT_ID -- digital IO *could* use timer event directly on this same node -- but it can also make use of the locally available tsplink two node[nodenum].digio.trigger[1].clear() node[nodenum].digio.trigger[1].mode = digio.TRIG_FALLING node[nodenum].digio.trigger[1].pulsewidth = 100e-6 --node[nodenum].digio.trigger[1].stimulus = node[nodenum].trigger.timer[1].EVENT_ID node[nodenum].digio.trigger[1].stimulus = node[nodenum].tsplink.trigger[2].EVENT_ID -- this remote node cannot "see" the timer events on other node -- this remote node must use the echoed event on tsplink two nodenum = 2 node[nodenum].digio.trigger[1].clear() node[nodenum].digio.trigger[1].mode = digio.TRIG_FALLING node[nodenum].digio.trigger[1].pulsewidth = 100e-6 node[nodenum].digio.trigger[1].stimulus = node[nodenum].tsplink.trigger[2].EVENT_ID -- issue an assert to get the party started node[1].tsplink.trigger[1].assert()
Scope shot of the digital IO pulses in attached document.
-
Use Digitize V to compute the RMS or Frequency of a Waveform
<p><p><a href="https://fortive.box.com/s/t3008ugydzhvlqzvcm6zjvixobm27bh8"><img alt="Digitize Voltage" src="https://fortive.box.com/s/t3008ugydzhvlqzvcm6zjvixobm27bh8" style="float:left;"></a></p> <p></p></p> <p></p> <p></p> <p><a href="https://fortive.box.com/s/t3008ugydzhvlqzvcm6zjvixobm27bh8"><img alt="Digitize V trial 2" src="https://fortive.box.com/s/t3008ugydzhvlqzvcm6zjvixobm27bh8" style="float:left;"></a></p>
-
RE: DAQ6510 OverflowHz/V Issue
Here is a little more general info on using the Digitize Voltage and the TSP scripting to extract frequency information from a measured voltage waveform.
For your waveform frequency and amplitude, configure the digitizer to obtain enough data to capture at least two peaks.
In attached document, is more info and a sample TSP script. -
RE: Using 2450 SMU in 4200 UTM
For others who might follow along:
2450 has USB, LAN and GPIB for remote command control.
Newer 4200A-SCS have VISA library available from KULT, so could use any of those command interfaces if we code at VISA API. Newer 4200A also have the GPIB interface API.
Older 4200-SCS do not have the VISA library. They have GPIB API for external instruments accessible from KULT and UTMs.
This sample code is written using GPIB API to do a voltage sweep on a model 2450 from a UTM in KITE or Clarius.
The 2450 requires the TSP command set and connected on GPIB to the 4200x.
In the attached files:
the *.c file is contain in a zip so that this forum tool permits it.
The PDF has instructions for installing it.