web
Вы не подключены к сети. Эта версия страницы доступна только для чтения.
close
  • RE: 2470 Triax Connectors

    The HV triax at rear of 2470 is deliberately different so that “conventional” triax cables from max 200V products cannot connect.

    The TRX-1100V-x cable is required for the HV triax on rear panel of 2470.

    Maybe what is unexpected:  The male triax connector on this cable is also compatible with “conventional” sized female triax barrels, feed through, etc.  However, do keep in mind the 1100V rating.
    There are various TRX-1100V-xxx accessories.  Also our legacy 237-xxx accessories are also rated for 1100v and are compatible with the HV triax cable.

    I advise you to consider the voltage rating of the probing setup and use 1100V components throughout.
  • RE: I have a question regarding the Keithely 6514 overhead time.

    Last page of Datasheet says 1200 readings per second into buffer at 0.01 nplc and with other things turned off.
    This gives some idea of the overhead.

    using trigger model to acquire N measurements into buffer will be most time determinant was to operate the 6514.
    Can the timestamps give you some info?

    Alternately, there is an analog signal on the preamp output.  You can sample that for exactly 1 second with scope if looking for area under current vs time curve for amp-hr info.
  • RE: How to speed-up triangle wave generation?

    The smu.source.sweeplinear() function is doing two things:
    - it builds a source configuration list containing voltage levels to source;   it also has source settings such as range and current limit, etc.
    - it builds a trigger model to loop through the source config list and get a measurement.

    To have control over the volt / sec ramp rate, need to modify the trigger model to make use of a timer to control the rate that it moves through the list.
    I'll upload some demo code, but keep in mind there are many opportunities for logical error pitfalls in it.
    For example, at faster desired ramp rates, need to make sure the measurements can keep up.
    But at 100msec at each source level, there is plenty of time.

    See attached.

     
  • RE: SMU 2612B - Firmware 4.x - LabTracer Issue

    Hello,
    I cannot think of any intended firmware differences in 4.x that would account for it.

    Can you capture a NI IO Trace during the attempt to use Labtracer with firmware 4.x and upload.
     
  • RE: Random measurement drop err. -113 Undefined Header

    On the old forum, user blegeyt gave helpful answer.

    The old post was for model 2400 but you have different model of product.

    Suggest:  use ni trace to capture the gpib bus traffic and see which exact command and error occurs.
    I suggest to verify root cause more fully.

    User blegeyt suggest to stop use of status byte reading and *opc, and instead *opc? and visa read with appropriate duration timeout.  He attributes the issue to usb to gpib interface or driver used by PC.
  • RE: DAQ6510 PLC COMM

    Yes.
    Typical way is to use raw sockets from the AB to talk to the instrument over LAN.
    Alternately, add a KTTI-rs232 communication module to the DAQ if the AB has COM port ability.

    This seems somewhat related:
    Using external equipment with AB PLC
  • RE: Hardware communication Error 5067 Keithley 2651A

    You might also want to contact David at the email in earlier post.

    Are you using any output off mode commands?
    Are you using the high capacitance mode?

    What is the Isc at 0V of your cell?
    What is the Voc at 0A?
  • RE: [4200A-SCS] KXCI recieve wrong command

    In KCON settings, your EOI is off.  Try turning that on in KCON for the gpib communication.
  • RE: Remote Restart SMU

    Thanks for clarifying.
    see if this post helps.  Same DST process for SMU or DAQ:
    https://my.tek.com/en/tektalk/switching-data-acq/8644f450-3c70-ee11-a81c-000d3a32997c

     
  • RE: How can I set "pulse mode" in keithley 2602B ?

    Needs an update:

    The code posted above will not take you into pulse region 4.
    These high watt pulse regions are in our Extended Operating Area and have firmware imposed max PW and Duty Cycle limits.

    To access the EOA regions, supply a higher voltage value for the trigger.source.limitv which will be applied only during the high watt pulse.
    When not sourcing the pulse level, the lower value set for the smuX.source.limitv applies.  The combo of smuX.source.leveli and smuX.source.limitv needs to be within the DC operating area of the SMU.

    These three lines in particular for accessing EOA region:

    smuX.measure.rangev = EOA_limit

    smuX.source.limitv = limitV

    smuX.trigger.source.limitv = EOA_limit

    I'll upload scope shot of 1.5A into 10Ω.

    Revised code:
    function Run_test(show_data, PulsePeriod, PulseWidth, MeasDelay)
    
    
       -- when passthrough is true, the timer will "tic" immediately,
       -- which for this application will start as soon as the
       -- trigger model is ARMED.  
       node[1].trigger.timer[1].delay = PulsePeriod
       node[1].trigger.timer[1].stimulus = node[1].smua.trigger.ARMED_EVENT_ID
       node[1].trigger.timer[1].count = 1  --PulseCount - 1
       node[1].trigger.timer[1].passthrough = true
       node[1].trigger.timer[1].clear()
       
       node[1].trigger.timer[2].delay = PulseWidth
       node[1].trigger.timer[2].stimulus = node[1].trigger.timer[1].EVENT_ID
       node[1].trigger.timer[2].count = 1        
       node[1].trigger.timer[2].passthrough = false
       node[1].trigger.timer[2].clear()
       
       node[1].trigger.timer[3].delay = MeasDelay
       node[1].trigger.timer[3].stimulus = node[1].trigger.timer[1].EVENT_ID
       node[1].trigger.timer[3].count = 1        
       node[1].trigger.timer[3].passthrough = false
       node[1].trigger.timer[3].clear()
    
    
        
        -- digital output trigger for external equipment
        -- configure digital IO line 1 to output a active LO/falling edge
        -- pulse at start of each current pulse
        digio.trigger[1].clear()
        digio.trigger[1].mode = digio.TRIG_FALLING  --digio.TRIG_RISINGM
        digio.trigger[1].pulsewidth = 100e-6
        digio.trigger[1].stimulus = node[1].trigger.timer[1].EVENT_ID  -- source stimulus too
        
            
       -- clear the reading buffers
       reset_buffers(node[1].smua)
         
     
    
        -- Turn the outputs on
        smua.source.output                    = smua.OUTPUT_ON
        
        -- after blue light state change, delay a little 
        delay(0.1)  
    
        -- Start the trigger model execution
        smua.trigger.initiate()   -- start this one last as all the other smus follow this leader
        
        -- Wait until the sweep has completed
        waitcomplete()
    
    --[[
    -- hot switch:
    -- to bleed off the trapped charge on PN junction capacitance and non-zero voltage
    -- briefly set the voltage compliance to small number
    restore_val = smua.source.limitv
    smua.source.limitv = 0.01
    delay(0.001)
    smua.source.limitv = restore_val
    ]]
    
        -- before blue light state change, delay a little 
        delay(0.1)
        -- Turn the output off
        smua.source.output                    = smua.OUTPUT_OFF
        
    
        -- IV data for the laser
        if show_data == true then
           -- Print the data back to the Console in tabular format
           print(" Voltage\tCurrent\tResistance")
           for x=1,smua.nvbuffer1.n do
              -- Voltage readings are in nvbuffer2.  Current readings are in nvbuffer1.
              print(smua.nvbuffer2[x], smua.nvbuffer1[x], smua.nvbuffer2[x]/smua.nvbuffer1[x] )
           end
        end
        
    
        
    end
    
    function reset_buffers(smu)
            -- Prepare the Reading Buffers
        smu.nvbuffer1.clear()
        smu.nvbuffer1.appendmode        = 1
        smu.nvbuffer1.collecttimestamps    = 1
        smu.nvbuffer2.clear()
        smu.nvbuffer2.appendmode        = 1
        smu.nvbuffer2.collecttimestamps    = 1
    
    end
    
    
    function config_smu_list_sweep(nodenum, smu, srcRange, srclist, limitV, EOA_limit, nplc, remoteSense)
        
        smu.source.func                    = smu.OUTPUT_DCAMPS
        if (remoteSense == true) then
            smu.sense                    = smu.SENSE_REMOTE
        else
            smu.sense                    = smu.SENSE_LOCAL
        end
        smu.source.autorangei            = smu.AUTORANGE_OFF
        smu.source.rangei                = srcRange
        smu.source.leveli                = 0
        smu.source.limitv                = limitV
    
        -- Disabling Auto-Ranging and Auto-Zero ensures accurate and consistent timing
        smu.measure.autozero            = smu.AUTOZERO_ONCE
        smu.measure.autorangev            = smu.AUTORANGE_OFF
        smu.measure.rangev                = EOA_limit
        smu.measure.nplc                = nplc
        -- A timer will be used to set the measure delay and synchronize the measurement
        -- between the SMUs so set the built in delay to 0.
        smu.measure.delay                = 0
        
        -- Configure SMU Trigger Model for Sweep
        --smu.trigger.source.lineari(start, stop, numPoints)
        smu.trigger.source.listi(srclist)
        smu.trigger.source.limitv        = EOA_limit
        smu.trigger.measure.action        = smu.ENABLE
        smu.trigger.measure.iv(smu.nvbuffer1, smu.nvbuffer2)
        smu.trigger.endpulse.action        = smu.SOURCE_IDLE
        smu.trigger.endsweep.action        = smu.SOURCE_IDLE
        smu.trigger.count                = table.getn(srclist)
        smu.trigger.arm.stimulus        = 0
        smu.trigger.source.stimulus        = trigger.timer[1].EVENT_ID
        smu.trigger.measure.stimulus    = trigger.timer[3].EVENT_ID
        smu.trigger.endpulse.stimulus    = trigger.timer[2].EVENT_ID
        smu.trigger.source.action        = smu.ENABLE
    end
    
    
    --  **************************************  MAIN PROGRAM BELOW *****************
    
    reset()
    errorqueue.clear()
    
    nodesFound = tsplink.reset(1)
    
    print("Nodes found = " .. nodesFound)
    
    --smua.interlock.enable = smua.DISABLE
    
    
    -- use of pulse region 4 of 2602B:  10A at up to 20V
    -- max 1.8msec pulse with 1% duty cycle
    local PulseWidth = 1.0e-3
    local PulsePeriod = PulseWidth / 0.01
    local MeasDelay = PulseWidth * 0.8
    local nplc = (PulseWidth - MeasDelay - 60e-6) * localnode.linefreq
    
    
    print("Computed NPLC value: "..tostring(nplc))
    
    if tostring(nplc) < "0.001" then
      nplc = 0.001   -- 16.7usec, minimum value allowed
      MeasDelay = 1e-6
    end -- if
    
    
    -- voltage ranges on 2602B:  100mV, 1V, 6V, 40V
    -- pick a limit value higher than expected I*DUT_R and
    -- puts you on appropriate range.
    local voltageLimit = 1
    local duringPulse_voltageLimit = 20
    
    -- ranges:  decades of 10.....100mA, 1A, 3A, 10A
    local mySrcRange = 3  
    mysrclist = { 1.5}  -- use this one for a single pulse
    --local numPulses = table.getn(mysrclist)
    
    
    
       
        --config_smu_list_sweep(nodenum, smu, srcRange, srclist, limitV, EOA_limit, nplc, remoteSense)
        config_smu_list_sweep(1, node[1].smua, mySrcRange, mysrclist, voltageLimit, duringPulse_voltageLimit, nplc, true)
    
    
    
       timer.reset()
       
       
       Run_test(true, PulsePeriod, PulseWidth, MeasDelay) -- pass true to print data back to console
       
       time = timer.measure.t()
       print("Time to run: "..time)
       print("****************************************")
       print("Current Source Range: "..node[1].smua.source.rangei)

    ​​​​​​​