RE: 2445 power supply failure, cap subs and optoisolator sub question
Hi Jason,
For the cap replacement question, I would suggest asking the TekScopes mailing list, Tektronix does not offer replacement caps.
For the optoisolator, it's important to know that Tek 9 digit part numbers have a system where the last 2 digits denote backwards compatible (but not necessarily forwards compatible) changes.
So, if your manual lists the optoisolator part number as 156-0885-00, 156-0885-05 should also work. -
RE: TDS 744A Attenuator Problem
Hi,
I've seen similar failures before, the contacts go bad and it leads to incorrect configuration (wrong amplitude) or AC coupling, like you see there.
It's just the relays of the attenuator network wearing out and going bad, a replacement would be ideal but if that's not possible, you could try very carefully cleaning them with alcohol or even just cycling them a few times. Tapping on them works as a temporary fix often times. -
RE: Oscilloscope Series 6
Hi Om,
waveform:data?
is not the correct command, instead you should use thecurve?
command.
Also, ASCII format is really inefficient, I suggest you instead use a binary format and then use query_binary_values() to read and decode it. -
RE: What are the TDS1000 and 2000 series Display Interface Protocols?
Hi Joel,
No, that's what I meant, there's a part number on the back of the LCD which would allow you to order a new one from a third party.
Unfortunately we generally do not share details about recent instrument design so I would not be able to share what protocols are used for the display. Hopefully you can still do your repair without that information! -
RE: Where can I find the file-format for the .sogram spectrogram data file XAML elements?
Hi Isaac,
Unfortunately the .sogram file format is not publicly documented, I'd suggest using SCPI to get the spectrogram data as we discussed previously and then saving it as your own custom format. -
RE: Oscilloscope Series 6
Hi Om,
Here's the latest version of TekVISA: https://www.tek.com/en/support/software/driver/tekvisa-connectivity-software-v420 -
RE: TDP1500 Specifications
Hi John,
The TDPs are active differential probes, the TAPs are our active single-ended probes, here's their product page: https://www.tek.com/en/products/oscilloscopes/oscilloscope-probes/low-voltage-probe-single-ended -
RE: Oscilloscope Series 6
Hi Om,
Do you have a signal that marks the start of your coincidence window? If so, you could set your scope to trigger on it, set the trigger position to 0% (all the way to the left) and then set your time/div to capture for the length you want (let's say you want a 50 ns window, you'd set the scope to 5 ns/div, or a little more) and set the trigger to norm mode, that way the scope will only capture when it receives a trigger signal as you defined it (in the trigger menu). -
RE: Hardcopy command to save PNG file didn't work for TBS2000B
Hi Jeremy,
The image format has to be set with theSAVe:IMAge:FILEFormat
command before transferring the image, as described on the programmer's manual. -
RE: Programmable oscilloscope MSO64B can't read HardCopy screenshot data,but MDO3054 can do it
Hi Suny,
I do, I've attached such an example with this post.