web
Bạn đang ngoại tuyến. Đây là phiên bản chỉ đọc của trang.
close
  • DC-DC Converter Efficiency test using two 2450 SourceMeter instruments

    From old forum:  https://forum.tek.com/viewtopic.php?f=263&t=143187

    In general, the problem in the code was how the voltage sourcing SMU at input to the DC-DC converter was programmed.

    Revised Python code for two 2450 attached.



     
  • RE: Problem Using DMM6500 with 2000-Scan Card

    Looks like this one found a work around:  https://forum.tek.com/viewtopic.php?f=617&t=142379
  • RE: DMM6500 Timer bug

    In addition to checking your firmware level, also take note that when you edit the Timer settings, an Enable Timer button is displayed.
    After the new delay value is entered, Press the Enable Timer button to apply those new settings to the timer.
    The Enable Timer button then disappears until you again have changed one of the timer settings.

    If you were operating from software control and using the timers, setting or editing timer values requires disabling the timer, set the various parameters and then enable it.

    Seems the touch screen GUI is detecting a change is made and awaiting for you to apply them with the Enable Timer button.

  • RE: AC Current Pulse with 2636 SMU

    Over GPIB, a program can send abort command to halt an executing script.
    For this older non-A and non-B model, the manual leaves me a little unclear on exact syntax.
    Try:  abort or abort() or smuX.abort()
    For the A and B versions, the smuX.abort() is the proper syntax.
  • RE: Keithley 590 C-V Analyzer: Applying the Calibration Capacitor Correction

    Curious which test freq does your 590 have?  100KHz or 1MHz or both?
    Which freq do you use?

    For the calibration capacitor topic, do you have the 5907 standard caps?  Or supplying your own caps?

    Are you connected through the switch when doing the cable compensation with the standard caps?
    You should be if you want the method to correct for it.

    You mention an unwanted offset of up to 20pF when connected through a switch.  Have you tried the zero method from section 3.10?

    Your 590 is likely to be quite old.  When was last time it was calibrated?
    How are you determining the value of the reference caps?

    In general, I’d try evaluating just the 590 with minimal cabling to the standard caps.  If successful, then start adding in the other aspects such as the switch, etc.
  • RE: Souring 0A current and get 1.5V on my doide

    For your finding1 topic:  when blue light is off, if the SMU is in default output off config, then it is behaving as a 0V source with ability to source or sink 1mA.
    This should result in 0V starting point.
    When you say the voltage was jumping around 0 to 2V, what was the condition of the SMU then?

    Any chance there was an open circuit due to probing issue?

    NOTE:  for PN junction, if using SMU to force current and forward bias the diode, it is very much possible to trap charge on the PN junction when the SMU returns to 0Amps and the blue light is still on, especially if using a low source range.
    Suppose you force current with a 3V limit.  When you return to 0Amps, the SMU has no trouble to allow as much as +/-3V.
    This should be fairly steady, but could decay due to leakage in the device, cables or allowed offset of the source range.
    You can bleed off the trapped charge by turning output off OR by temporarily setting the voltage limit to smaller value.
  • RE: AC Current Pulse with 2636 SMU

    the sync_in and sync_out parameters of the Config functions will need to be used.

    Wondering if you can just config each SMU to make use of digital line 1.  And then use the digio.trigger[1].assert() to satisfy the digital event.

    Also, check your firmware and update to the final release (1.4.2) for these non-A and non-B vintage models.
    firmware for 2600 nonA and nonB
  • RE: 2602B TSP command equivalent to the EXIT (LOCAL) front panel button

    If you are using VISA to exchange messages with the instrument, when you clear and then close the session, the REM indicator on the 2600B will go out and the auto initiated local mode readings will resume.
     

    import pyvisa
    import time
    
    
    resource_mgr = pyvisa.ResourceManager()
    #optional print available VISA resources on this computer
    resource_list = resource_mgr.list_resources()
    for resource in resource_list:
        print(resource)
    
    try:
         #get a session/handle for one resource
         instrument_resource_string = resource_list[0]  #"USB0::0x05E6::0x2602::4484585::INSTR"
         #"TCPIP0::192.168.1.50::inst0::INSTR"  #
         my_instr = resource_mgr.open_resource(instrument_resource_string)
    except(visa.VisaIOError):
         #did not connect.....
         print("Whoops, something went wrong")
    
    my_instr.write("*IDN?\n")
    print("*****")
    print(my_instr.read())
    
    #put instrument back to local and close connection
    my_instr.clear()
    my_instr.close()
    resource_mgr.close()


     

  • RE: Souring 0A current and get 1.5V on my doide

    Looking at the source spec for the 1.5Amp range, you could have as much as +/-4mA of current and still be within spec for source value of 0Amps.
    In contrast, when using 1uA range to source 0A, the error budget drops to +/-700pA.

    If wanting the SMU to behave like a volt meter, in general, use the most sensitive range available to source the 0Amps.  Set the corresponding voltage compliance limit to a value larger than the DUT can produce.  Measure V.


    For your finding1, what was the state of the SMU before landing the probes?  Was it already in Force I, Measure V?  Was blue light on?
  • RE: Series 2200 - USBTMC Error - Device not found

    Hi Greg,

    I doubt if this will help, but just in case I wanted to mention to double check the Windows Device Manager.
    If Windows has mistake in how it categorized the USB connected instrument, then it would prevent proper association to the USB Test and Measurement device type, the IVI components, showing up in NI-MAX, etc.
    Sounds like your specific unit is not being enumerated at all by Windows.
    However, if it was instead being detected and assigned to "Other devices" or something else, this could account for the trouble.
    Sometimes, if no driver is present when the usb is first plugged in, an incorrect categorization is made and Windows will just repeat it even after a driver has been installed.
    In that case, delete the incorrect entry from Device Manager and then un-plug and re-plug the USB connection and see if it will correct the association to USBTMC.