  • RE: How to detect end of list sweep in a python (pyvisa) script?

    If the instrument is busy carrying out the sweep, then the VISA write of the "*OPC?" command is not processed until the sweep completes.  This is why the VISA read will timeout.

    An alternate approach is to use VISA's ReadSTB() to obtain the value of the status byte or status.condition register.

    I put an elaboration and Python sample in the attached document.
  • RE: 2611 SMU (NON A, NON B) Programming Reference

    Hello,
    The PDF is too large for attaching, so here is link to Box:

    Reference Manual for the original 2600 models
  • RE: Issues with Connecting Keithley 2450 to PC due to IP Address Changing

    The LAN communication can be applied from your router (DHCP) or selected manually.

    Sounds like you have been using DHCP and the router is periodically giving new address?

    Perhaps you can reserve an address in the router setup and manually configure the instrument to use it?
  • RE: SMU 2400 Series ErrorCode

    Appendix B-2 in the 2400 Series User Manual has the error codes.
    NOTE:  this is not the touch screen models (2450, etc).

    User Manual for 2400 Series SourceMeter
  • RE: Multiplexing SMU 2450

    Some general comments:

    For switching pA levels, consider the offset current spec of the switch.

    For parallel reliability testing, consider using a power supply to apply voltage bias to N devices.
    And then switch in a sensitive amp meter one by one.
    When not measuring the current, the switching scheme needs to maintain pathway completion for the current to flow.
    Model 3761 switch is one example.
  • RE: Linking two 2450 SMU

    KickStart does not use tsp-link for communication with the SMU.
    KickStart needs a computer connection to each SMU (lan, usb or gpib).

    If you additionally have tsp-link between the instruments, KickStart can use that for their triggering/coordination.
     
  • RE: How to upgrade firmware 3.4.2 to 4.0.3 [2600B]

    The 4.x firmware of 2600B series apply only to units shipped from factory with a 4.x level.

    If your 2600B has 3.x level, then 3.4.2 from Tek.com would be the most recent version.
  • RE: Simultaneous sine waves on SMU 2602 channels A & B

    Here is a lightly edited version of your code:
    local _startV -- Used by script to track the starting voltage
    local _srcRate = 4000 -- Source Update Rate used by the script Sets the source update rate (pts/sec) used by the script. 
                          --Do not set higher than 8,000 if your waveform will have polarity changes 0V crossings)
        
    function SetupAWG(startV, rangeV, limitI, wfrmTbl, remoteSense, trigLineIn)
        -- Do some parameter checks
        --=========================
        if startV == nil then startV = 0 end
        if remoteSense ~= true then remoteSense = false end
        if type(trigLineIn) == "number" then
            trigLineIn = math.floor(trigLineIn)
            if trigLineIn < 0 or trigLineIn > 14 then
                return true, "Error: Selected trigger line is not valid.  trigLineIn must be a number between 0 and 14 or nil."
            end
        elseif trigLineIn ~= nil then
            return true,"Error: Invalid parameter trigLineIn.  trigLineIn must be a number between 0 and 14 or nil."
        end
        _startV = startV
        
        --wfrmTbl = {0.078459,0.156434,0.233445,0.309017,0.382683,0.45399,0.522499,0.587785,0.649448,0.707107,0.760406,0.809017,0.85264,0.891007,0.92388,0.951057,0.97237,0.987688,0.996917,1,0.996917,0.987688,0.97237,0.951057,0.92388,0.891007,0.85264,0.809017,0.760406,0.707107,0.649448,0.587785,0.522499,0.45399,0.382683,0.309017,0.233445,0.156434,0.078459,0,-0.078459,-0.156434,-0.233445,-0.309017,-0.382683,-0.45399,-0.522499,-0.587785,-0.649448,-0.707107,-0.760406,-0.809017,-0.85264,-0.891007,-0.92388,-0.951057,-0.97237,-0.987688,-0.996917,-1,-0.996917,-0.987688,-0.97237,-0.951057,-0.92388,-0.891007,-0.85264,-0.809017,-0.760406,-0.707107,-0.649448,-0.587785,-0.522499,-0.45399,-0.382683,-0.309017,-0.233445,-0.156434,-0.078459,0}      
       
       -- compute the sine wave
        pts_per_cycle = 72
        wfrmTbl = {}
        for i=1, pts_per_cycle do
            wfrmTbl[i] = 1 * math.sin(i * 2 * math.pi/pts_per_cycle)
        end
        
        -- use some phase shift for smub sine wave
        wfrmTbl_smub = {}
        for i=1, pts_per_cycle do
            wfrmTbl_smub[i] = 1 * math.sin(i * 2 * math.pi/pts_per_cycle + math.rad(90))
        end
        
        
        
            
        -- Setup the SMU for arb waveform output
        --======================================
        reset()
        smua.reset()
        smub.reset()
        
        -- digital output trigger for external equipment
        -- configure digital IO line 1 to output a active LO/falling edge
        -- pulse at start of each current pulse
        digio.trigger[1].clear()
        digio.trigger[1].mode = digio.TRIG_FALLING  
        digio.trigger[1].pulsewidth = 10e-6
        digio.trigger[1].stimulus = trigger.timer[1].EVENT_ID
        
        smua.source.func                    = smua.OUTPUT_DCVOLTS
        smub.source.func                    = smub.OUTPUT_DCVOLTS
        if remoteSense == true then
            smua.sense                        = smua.SENSE_REMOTE
            smub.sense                        = smub.SENSE_REMOTE
        else
            smua.sense                        = smua.SENSE_LOCAL
            smub.sense                        = smub.SENSE_LOCAL
        end
        smua.source.autorangev            = smua.AUTORANGE_OFF
        smua.source.autorangei            = smua.AUTORANGE_OFF
        smua.source.rangev                = rangeV
        smua.source.levelv                = startV
        smua.source.limiti                = limitI
        smua.source.delay                = 0
        smua.source.settling            = smua.SETTLE_FAST_POLARITY
        
           
        smub.source.autorangev            = smub.AUTORANGE_OFF
        smub.source.autorangei            = smub.AUTORANGE_OFF
        smub.source.rangev                = rangeV
        smub.source.levelv                = startV
        smub.source.limiti                = limitI
        smub.source.delay                = 0
        smub.source.settling            = smub.SETTLE_FAST_POLARITY
        
        
        -- Configure the Trigger Model
        --============================
        -- Timer 1 controls the time per point table phase A
        trigger.timer[1].delay            = 1 / _srcRate
        trigger.timer[1].count            = table.getn(wfrmTbl) > 1 and table.getn(wfrmTbl) - 1 or 1
        if trigLineIn == nil then
            -- Immediate
            trigger.timer[1].stimulus    = smua.trigger.ARMED_EVENT_ID
            
        elseif trigLineIn == 0 then
            -- Front panel TRIG button
            display.trigger.clear()
            trigger.timer[1].stimulus    = display.trigger.EVENT_ID
        else
            -- Digio Trigger
            digio.trigger[trigLineIn].clear()
            digio.trigger[trigLineIn].mode = digio.TRIG_EITHER
            trigger.timer[1].stimulus    = digio.trigger[trigLineIn].EVENT_ID
        end
        trigger.timer[1].passthrough    = true
        
        
        
    
        -- Configure SMU Trigger Model for arb waveform output
        smua.trigger.source.listv(wfrmTbl)
        smua.trigger.source.limiti        = limitI
        smua.trigger.measure.action        = smua.DISABLE
        smua.trigger.endpulse.action    = smua.SOURCE_HOLD
        smua.trigger.endsweep.action    = smua.SOURCE_HOLD  
        smua.trigger.count                = table.getn(wfrmTbl)
        smua.trigger.arm.count            = 1
        smua.trigger.arm.stimulus        = 0
        smua.trigger.source.stimulus    = trigger.timer[1].EVENT_ID
        smua.trigger.measure.stimulus    = 0
        smua.trigger.endpulse.stimulus    = 0
        smua.trigger.source.action        = smua.ENABLE
        
            
        smub.trigger.source.listv(wfrmTbl_smub)
        smub.trigger.source.limiti        = limitI
        smub.trigger.measure.action        = smub.DISABLE
        smub.trigger.endpulse.action    = smub.SOURCE_HOLD
        smub.trigger.endsweep.action    = smub.SOURCE_HOLD
        smub.trigger.count                = table.getn(wfrmTbl)
        smub.trigger.arm.count            = 1
        smub.trigger.arm.stimulus        = 0
        smub.trigger.source.stimulus    = trigger.timer[1].EVENT_ID  -- ALC
        smub.trigger.measure.stimulus    = 0
        smub.trigger.endpulse.stimulus    = 0
        smub.trigger.source.action        = smub.ENABLE
        
        --==============================
        -- End Trigger Model Configuration
        
        if errorqueue.count > 0 then
            return true,"Error occured during setup.  Please check that your parameters are valid."
        else
            return false,"No error."
        end
    end
    
    
    function RunAWG(numCycles)
        if numCycles == nil or numCycles < 0 then
            numCycles = 1
        end
        
        -- Set the number of cycles to output
        smua.trigger.arm.count = numCycles
        smub.trigger.arm.count = numCycles
        -- Turn output on
        smua.source.output = smua.OUTPUT_ON
        smub.source.output = smub.OUTPUT_ON
        -- Start the trigger model execution
        smub.trigger.initiate()  -- start this first
        smua.trigger.initiate()
        
        if errorqueue.count > 0 then
            return true,"Error occurred.  See error queue for details."
        else
            return false,"No error."
        end
    end
    
    
    --[[ Name: StopAWG()
    
        Usage: err,msg = StopAWG()
    
        Description:
            This function stops the waveform output and turns the SMU
            output off.
    --]]
    function StopAWG()
        smua.abort()
        smua.source.output = 0
        smua.source.levelv = _startV
        
        smub.abort()
        smub.source.output = 0
        smub.source.levelv = _startV
        
        if errorqueue.count > 0 then
            return true,"Error occured.  See error queue for details."
        else
            return false,"No error."
        end
    end

    To run it:
    --SetupAWG(startV, rangeV, limitI, wfrmTbl, remoteSense, trigLineIn)
    SetupAWG(0, 20, 0.1, wfrmC, false, nil)
    
    --RunAWG(numCycles)
    RunAWG(8)

    See scope shot attached.
  • RE: Simultaneous sine waves on SMU 2602 channels A & B

    Hello,

    You are almost there.

    To accomplish the phase shift, were you thinking to use a different timer for each SMU?
    Instead, I would use a single timer to provide the smuX.trigger.source.stimulus to each SMU channel.

    Rather than providing a list of source levels, I recommend computing them. in the TSP or Lua code.
    You can account for the phase shift in the computed source levels.
       
       -- compute the sine wave
        pts_per_cycle = 72
        wfrmTbl = {}
        for i=1, pts_per_cycle do
            wfrmTbl[i] = 1 * math.sin(i * 2 * math.pi/pts_per_cycle)
        end
        
        -- use some phase shift for smub sine wave
        wfrmTbl_smub = {}
        for i=1, pts_per_cycle do
            wfrmTbl_smub[i] = 1 * math.sin(i * 2 * math.pi/pts_per_cycle + math.rad(90))
        end
  • RE: 2636B: V pulse and transient current measurement

    You second task can be implemented with the asynchronous trigger model.

    Old forum showing the feature:
    asynchronous sampling from the source timing - capture transients

    I will upload a simple LabVIEW sample for loading scripts/functions and then making use of the functions from VIs.