  • RE: Modernizing a legacy 2410 SMU

    In addition to the 2450 related accessory, the 1100V model 2470 has similar one:  2470-1100V-BAN.

    I see no reason not to operate them in reverse, e.g. plug the banana into rear of 2410 to present triax.

    Take note on how the LO and Sense LO are handled if intending to use 4-wire connections.

    Info about 2470-1100V Accessories

    The male triax ends are mechanically compatible with the rear panel HV triax connectors on model 2470.
    But they are also compatible with our more legacy triax such as the 237 related barrels, tees, 7078-TRX-x, CS-630, etc.
    You may need some 237-TRX-BAR barrel connectors.

     
  • RE: SMU IV Connections

    So, your terminal 4 is not at GND potential?<br> How big might the voltage at terminal 4 become?<br> <br> I do think you will need a floating amp meter where the LO is not shared with other SMU in 4200.<br> An external 2400 or 2600….. or even Amps terminals of a DMM depending on the values of current.<br> <br> Can you use known small resistor between terminal 3 and 4. &#160;Use a SMU at each terminal too.<br> Both SMU force 0 Amps and measure V. &#160;Your current is (V3-V4)/R<br> <br> A third SMU sweeps V at terminal 1.<br> Terminal 2: &#160;GNDU or use a SMU to force 0V and measure the current flowing into the SMU. &#160;Be sure to set the current limit high enough for all the current that will arrive.
  • RE: Stop condition on overlapped sweeps

    See if this helps. &#160;The code should apply to any 2600A or 2600B.<br> https://forum.tek.com/viewtopic.php?f=224&amp;t=142633<br> <br> &#160;
  • RE: Strange IV curve of solar cell measurement by 2651A

    Try turning on the high capacitance mode.<br> smua.source.highc = 1
  • RE: Update rate of K2450

    Hello,
    I'm not sure if a bug or feature.
    But I have opened a formal issue in our quality tracking system.
    My expectation was that we should be able to turn off the delays even if it is likely to wreck your data quality.
    As for the 300msec that is imposed for use of the most sensitive ranges, it is a reasonable number.
    If operating in the nA levels, is reasonable to assume there are GΩ of resistance.
    As for estimating the capacitance, is easy to envision 100pF from a few feet of cable.
    At 1GΩ and 100pF, the RC time constant is 100msec.
    For settled readings, is not uncommon to wait 5 time constants.  It really depends on the setup and particular goals.
    Keep in mind, when the voltage steps from one value to the next, a displacement current is generated in proportion to the C and the step size and edge speed ( I = C * dV/dt).
    The delay is targeting enough time to have a reasonably steady state measurement.
    I will mention our 2600B series have more flexible trigger models. By this I mean it can measure Asynchronously from the sourcing action, so you can capture the transient displacement currents as the voltage is stepped.   
    Andrea
  • RE: Update rate of K2450

    What is your firmware level?&#160; I've been using 1.7.12b on my 2450 SourceMeter.<br> <br> If source delay is zero and auto delay is off, my understanding is that there should not be any current range dependent timing.<br> Of course, if using sensitive ranges where settling time is likely to be required given the RC time constants, operating the instrument this way can easily result in unsettled measurements.<br> <br> I used the same code but add a command to set the smu.measure.count equal to three, and use the relativetimestamps differences to get an idea about how fast the instrument is going.<br> So readings 1 through 3 are at source level 1;&#160; readings 4 through 6 are source level 2, etc.<br> <br> Regardless of range, the delta in timestamps for readings obtained at same source level is about 320usec.<br> But when comparing the timestamps from before and after a source level change, we have to expect some larger amount of time for the sourcing action to take place.<br> However, I also see that the amount of time in those time stamps does vary with the current measure range setting and for the 10nA and 100nA ranges it is about 300msec.&#160; But for 10uA and higher ranges, the delta timestamp is about 900usec.<br> <br> Timestamp delta for after src change to before src change:<br> print(my_instr.query(&quot;print(defbuffer1.relativetimestamps[4] - defbuffer1.relativetimestamps[3] )&quot;))<br> <br> <img alt="" src="https://fortive.box.com/shared/static/l5ikdfyii7iel6yjzbjd03g5d06wwj5d.png">
  • RE: Partial waveforms being generated

    Adding additional information: &#160;If the SEQ option license is enabled, then the start of the waveform can be controlled by a WAIT event.<br> For SCPI wait, when your software sends *TRG, the SCPI wait event is satisfied and the defined waveform will commence.<br> <br> <a href="https://imgur.com/a/Eyq2lQA"><img alt="" src="https://imgur.com/a/Eyq2lQA" style="float:left;"></a><a href="https://imgur.com/a/Eyq2lQA"><img alt="" src="https://imgur.com/a/Eyq2lQA"></a>
  • RE: How to use the status model on a 3706A

    <p>When you say you cannot reset the status registers, can you elaborate?</p> <p>Is it that the status.condition register still has some bits at logic 1 but you expect 0?&#160; Which ones/what is the value of the register?</p> <p>For example, if data is still sitting in the output queue (from print or printbuffer), the MAV bit will be set.</p> <p>Either VISA read the info or send a inst.clear() to clear old data in output queue and that will reset the MAV bit to 0.</p> <p>The status reset affects the enable and other writable registers.&#160; The read only condition registers are not affected by status reset.</p>
  • RE: How to use the status model on a 3706A

    Hello Lars,<br> <br> See if the attached document helps.<br> <br> &#160;
  • RE: How to automate sweeping every 10 min

    At subsite level in the project, go to subsite setup and enable Cycle Mode.<br> Then if you highlight the subsite when clicking the Run button, all enabled tests under that level will be executed once per cycle.&#160; The data will appear at each test but rather than Append2, Append3, etc.&#160; They will be cycle1, cycle2, etc.<br> <br> How to also have some pause for 10 minutes for each cycle?<br> Do you have the compiler a the KULT tool is functional on your 4200?<br> If yes, you could make a simple routine that waits for the desired duration before returning control back to KITE for the cycles to continue.<br> <br> NOTE:&#160; on the new 4200A-SCS, the compiler is always there.&#160; Older 4200-SCS, depending on vintage, needed it installed as an option.&#160; Open the KULT tool from desktop icon and see if it complains or not.