The 2600-PCT-2B contains two SourceMeter models:  2651A and 2636B. Many (not all) of the specs you seek will be in the datasheet for the individual SMUs. I encourage you to contact your closest sales office and discuss your application and concerns.
-
Is the sample in any sort of shielded test fixture?<br>
<br>
Double check your connections.  Are you making use of meter-connect to connect LO of Amps meter to the LO of the V-Source?<br>
<br>
When was your electrometer last calibrated? Perhaps try with an axial leaded high ohms device as practice and to raise confidence in function of the instrument.<br>
<br>
There are many instructional videos on the 6517B web page.<br>
Try this one:  https://www.tek.com/en/video/how-to/how-to-make-a-resistance-measurement-using-the-model-6517b-electrometer-with-auto-and-manual-modes
-
Check out this application note pertaining to our 4200A-SCS and 1/f noise measurements:<br>
<a href="https://www.tek.com/en/documents/application-note/flicker-noise-measurements-using-the-4200a-scs-parameter-analyzer">1/F Noise Measurement with 4200A-SCS</a>
-
Here is some Python code that might more in line with what you are trying to do.<br>
I will upload an image of the outputs I get.  I have been feeding a 500Hz sine wave to the DMM7510.
<pre class="linenums prettyprint">debug = 1
NUM_SAMPLES = 100
my_instr.write('*RST')
my_instr.write(':TRACe:MAKE "voltDigBuffer",' + str(NUM_SAMPLES))
my_instr.write(':DIG:FUNC "VOLTage"')
my_instr.write(':SENS:DIG:VOLT:INP AUTO')
my_instr.write(':SENS:DIG:VOLT:RANG 10')
my_instr.write(':SENS:DIG:VOLT:SRATE 5000')   #5KHz sample rate
my_instr.write(':SENS:DIG:VOLT:APER AUTO')
my_instr.write(':SENS:DIG:COUN ' + str(NUM_SAMPLES))
my_instr.write(':TRAC:POIN ' + str(NUM_SAMPLES))
my_instr.write(':TRAC:CLE')
my_instr.write(':TRAC:TRIG:DIG "voltDigBuffer"')
# wait for done before asking for the data
time.sleep(1)  #pause before asking for data
if debug == 1:
    my_instr.write(':TRAC:DATA? 95, 100, "voltDigBuffer", REL, READ')
    print("****** response to last five points query *************")
    print(my_instr.read())
#split, parse, etc.
"""
raw_data will be comma delimted string of 
timestamp, reading, timestamp, reading,... etc.
"""
#ask for all the data
#TODO:  adjust for NUM_SAMPLES
raw_data = my_instr.query(':TRAC:DATA? 1, 100, "voltDigBuffer", REL, READ')
raw_data_array = raw_data.split(',')
timestamps = []
Digitized_V = []
# use step of 2 because there are two elements per reading
for i in range(0, len(raw_data_array), 2):
    if len(raw_data_array[i]) > 0:
        timestamps.append(float(raw_data_array[i]))
        Digitized_V.append(float(raw_data_array[i+1]))
if debug == 1:
    print("******* Timestamps *******************")        
    print(timestamps)
    print("******** Voltage ******************")
    print(Digitized_V)
    print("**************************")
    print()
#graph it
plt.autoscale(True, True, True)
#plt.axis([0, 0.2, 0, 50e-6])
plt.plot(timestamps, Digitized_V, 'c--')
plt.show()
</pre>
-
In general, when the command you send ends with the question mark (a query type command), you should immediately follow with a read to pull the response to the query command off the output queue.<br>
<br>
You are sending more than one command in a row.  The first time you read, you will get the response from the first query command, not from the most recent.<br>
<br>
I'll attach a screen image of your code and the outputs but using a read after each query command.
-
Can you successfully ping the IP address of the instrument?<br>
What's the IP address of the computer from which you are trying this?<br>
<br>
From PuTTY, is straightforward to get a Telnet session on port 23.  See attached document.
-
The TP3005DM is not a model I recognize.  A google search shows it as made by company “TekPower” not Tektronix.<br>
<br>
Consider our various single channel models from the 2200 series.<br>
https://www.tek.com/en/products/keithley/dc-power-supplies/2200-dc-power-supply<br>
 
-
Are you moving the scope card between the two systems?  Or do they both have a scope card and just the one running KITE 9.1 is not working?<br>
<br>
If you are attempting to move the card between the two systems, it will not work.  The expected cards of the configuration is not designed to be modified in the field.  So if the KITE 9.1 system never had a scope card, it will not be accepted by the KCON config.<br>
<br>
Please clarify.
-
Ayoub,<br>
<br>
Your post seem off topic, so I created a different thread for managing 4200A-CVIV switch.  <br>
https://my.tek.com/tektalk/semi-para-analyzer/7654133b-38da-ed11-a81c-00224809317b<br>
<br>
Andrea<br>
 
-
In Clarius software, be sure to make use of an Action to route the SMU or CVU to the output.<br>
Execute the Action before you attempt to perform the SMU or CVU measurement.<br>
In the Select ribbon, search for a project using "CVIV" as the search term.<br>
Those projects will include the Action steps for configuring the CVIV prior to a measurement.