• RE: SMU 2400 Series ErrorCode

    Appendix B-2 in the 2400 Series User Manual has the error codes.
    NOTE:  this is not the touch screen models (2450, etc).

    User Manual for 2400 Series SourceMeter
  • RE: Multiplexing SMU 2450

    Some general comments:

    For switching pA levels, consider the offset current spec of the switch.

    For parallel reliability testing, consider using a power supply to apply voltage bias to N devices.
    And then switch in a sensitive amp meter one by one.
    When not measuring the current, the switching scheme needs to maintain pathway completion for the current to flow.
    Model 3761 switch is one example.
  • RE: Linking two 2450 SMU

    KickStart does not use tsp-link for communication with the SMU.
    KickStart needs a computer connection to each SMU (lan, usb or gpib).

    If you additionally have tsp-link between the instruments, KickStart can use that for their triggering/coordination.
     
  • RE: How to upgrade firmware 3.4.2 to 4.0.3 [2600B]

    The 4.x firmware of 2600B series apply only to units shipped from factory with a 4.x level.

    If your 2600B has 3.x level, then 3.4.2 from Tek.com would be the most recent version.
  • RE: Simultaneous sine waves on SMU 2602 channels A & B

    Here is a lightly edited version of your code:
    local _startV -- Used by script to track the starting voltage
    local _srcRate = 4000 -- Source Update Rate used by the script Sets the source update rate (pts/sec) used by the script. 
                          --Do not set higher than 8,000 if your waveform will have polarity changes 0V crossings)
        
    function SetupAWG(startV, rangeV, limitI, wfrmTbl, remoteSense, trigLineIn)
        -- Do some parameter checks
        --=========================
        if startV == nil then startV = 0 end
        if remoteSense ~= true then remoteSense = false end
        if type(trigLineIn) == "number" then
            trigLineIn = math.floor(trigLineIn)
            if trigLineIn < 0 or trigLineIn > 14 then
                return true, "Error: Selected trigger line is not valid.  trigLineIn must be a number between 0 and 14 or nil."
            end
        elseif trigLineIn ~= nil then
            return true,"Error: Invalid parameter trigLineIn.  trigLineIn must be a number between 0 and 14 or nil."
        end
        _startV = startV
        
        --wfrmTbl = {0.078459,0.156434,0.233445,0.309017,0.382683,0.45399,0.522499,0.587785,0.649448,0.707107,0.760406,0.809017,0.85264,0.891007,0.92388,0.951057,0.97237,0.987688,0.996917,1,0.996917,0.987688,0.97237,0.951057,0.92388,0.891007,0.85264,0.809017,0.760406,0.707107,0.649448,0.587785,0.522499,0.45399,0.382683,0.309017,0.233445,0.156434,0.078459,0,-0.078459,-0.156434,-0.233445,-0.309017,-0.382683,-0.45399,-0.522499,-0.587785,-0.649448,-0.707107,-0.760406,-0.809017,-0.85264,-0.891007,-0.92388,-0.951057,-0.97237,-0.987688,-0.996917,-1,-0.996917,-0.987688,-0.97237,-0.951057,-0.92388,-0.891007,-0.85264,-0.809017,-0.760406,-0.707107,-0.649448,-0.587785,-0.522499,-0.45399,-0.382683,-0.309017,-0.233445,-0.156434,-0.078459,0}      
       
       -- compute the sine wave
        pts_per_cycle = 72
        wfrmTbl = {}
        for i=1, pts_per_cycle do
            wfrmTbl[i] = 1 * math.sin(i * 2 * math.pi/pts_per_cycle)
        end
        
        -- use some phase shift for smub sine wave
        wfrmTbl_smub = {}
        for i=1, pts_per_cycle do
            wfrmTbl_smub[i] = 1 * math.sin(i * 2 * math.pi/pts_per_cycle + math.rad(90))
        end
        
        
        
            
        -- Setup the SMU for arb waveform output
        --======================================
        reset()
        smua.reset()
        smub.reset()
        
        -- digital output trigger for external equipment
        -- configure digital IO line 1 to output a active LO/falling edge
        -- pulse at start of each current pulse
        digio.trigger[1].clear()
        digio.trigger[1].mode = digio.TRIG_FALLING  
        digio.trigger[1].pulsewidth = 10e-6
        digio.trigger[1].stimulus = trigger.timer[1].EVENT_ID
        
        smua.source.func                    = smua.OUTPUT_DCVOLTS
        smub.source.func                    = smub.OUTPUT_DCVOLTS
        if remoteSense == true then
            smua.sense                        = smua.SENSE_REMOTE
            smub.sense                        = smub.SENSE_REMOTE
        else
            smua.sense                        = smua.SENSE_LOCAL
            smub.sense                        = smub.SENSE_LOCAL
        end
        smua.source.autorangev            = smua.AUTORANGE_OFF
        smua.source.autorangei            = smua.AUTORANGE_OFF
        smua.source.rangev                = rangeV
        smua.source.levelv                = startV
        smua.source.limiti                = limitI
        smua.source.delay                = 0
        smua.source.settling            = smua.SETTLE_FAST_POLARITY
        
           
        smub.source.autorangev            = smub.AUTORANGE_OFF
        smub.source.autorangei            = smub.AUTORANGE_OFF
        smub.source.rangev                = rangeV
        smub.source.levelv                = startV
        smub.source.limiti                = limitI
        smub.source.delay                = 0
        smub.source.settling            = smub.SETTLE_FAST_POLARITY
        
        
        -- Configure the Trigger Model
        --============================
        -- Timer 1 controls the time per point table phase A
        trigger.timer[1].delay            = 1 / _srcRate
        trigger.timer[1].count            = table.getn(wfrmTbl) > 1 and table.getn(wfrmTbl) - 1 or 1
        if trigLineIn == nil then
            -- Immediate
            trigger.timer[1].stimulus    = smua.trigger.ARMED_EVENT_ID
            
        elseif trigLineIn == 0 then
            -- Front panel TRIG button
            display.trigger.clear()
            trigger.timer[1].stimulus    = display.trigger.EVENT_ID
        else
            -- Digio Trigger
            digio.trigger[trigLineIn].clear()
            digio.trigger[trigLineIn].mode = digio.TRIG_EITHER
            trigger.timer[1].stimulus    = digio.trigger[trigLineIn].EVENT_ID
        end
        trigger.timer[1].passthrough    = true
        
        
        
    
        -- Configure SMU Trigger Model for arb waveform output
        smua.trigger.source.listv(wfrmTbl)
        smua.trigger.source.limiti        = limitI
        smua.trigger.measure.action        = smua.DISABLE
        smua.trigger.endpulse.action    = smua.SOURCE_HOLD
        smua.trigger.endsweep.action    = smua.SOURCE_HOLD  
        smua.trigger.count                = table.getn(wfrmTbl)
        smua.trigger.arm.count            = 1
        smua.trigger.arm.stimulus        = 0
        smua.trigger.source.stimulus    = trigger.timer[1].EVENT_ID
        smua.trigger.measure.stimulus    = 0
        smua.trigger.endpulse.stimulus    = 0
        smua.trigger.source.action        = smua.ENABLE
        
            
        smub.trigger.source.listv(wfrmTbl_smub)
        smub.trigger.source.limiti        = limitI
        smub.trigger.measure.action        = smub.DISABLE
        smub.trigger.endpulse.action    = smub.SOURCE_HOLD
        smub.trigger.endsweep.action    = smub.SOURCE_HOLD
        smub.trigger.count                = table.getn(wfrmTbl)
        smub.trigger.arm.count            = 1
        smub.trigger.arm.stimulus        = 0
        smub.trigger.source.stimulus    = trigger.timer[1].EVENT_ID  -- ALC
        smub.trigger.measure.stimulus    = 0
        smub.trigger.endpulse.stimulus    = 0
        smub.trigger.source.action        = smub.ENABLE
        
        --==============================
        -- End Trigger Model Configuration
        
        if errorqueue.count > 0 then
            return true,"Error occured during setup.  Please check that your parameters are valid."
        else
            return false,"No error."
        end
    end
    
    
    function RunAWG(numCycles)
        if numCycles == nil or numCycles < 0 then
            numCycles = 1
        end
        
        -- Set the number of cycles to output
        smua.trigger.arm.count = numCycles
        smub.trigger.arm.count = numCycles
        -- Turn output on
        smua.source.output = smua.OUTPUT_ON
        smub.source.output = smub.OUTPUT_ON
        -- Start the trigger model execution
        smub.trigger.initiate()  -- start this first
        smua.trigger.initiate()
        
        if errorqueue.count > 0 then
            return true,"Error occurred.  See error queue for details."
        else
            return false,"No error."
        end
    end
    
    
    --[[ Name: StopAWG()
    
        Usage: err,msg = StopAWG()
    
        Description:
            This function stops the waveform output and turns the SMU
            output off.
    --]]
    function StopAWG()
        smua.abort()
        smua.source.output = 0
        smua.source.levelv = _startV
        
        smub.abort()
        smub.source.output = 0
        smub.source.levelv = _startV
        
        if errorqueue.count > 0 then
            return true,"Error occured.  See error queue for details."
        else
            return false,"No error."
        end
    end

    To run it:
    --SetupAWG(startV, rangeV, limitI, wfrmTbl, remoteSense, trigLineIn)
    SetupAWG(0, 20, 0.1, wfrmC, false, nil)
    
    --RunAWG(numCycles)
    RunAWG(8)

    See scope shot attached.
  • RE: Simultaneous sine waves on SMU 2602 channels A & B

    Hello,

    You are almost there.

    To accomplish the phase shift, were you thinking to use a different timer for each SMU?
    Instead, I would use a single timer to provide the smuX.trigger.source.stimulus to each SMU channel.

    Rather than providing a list of source levels, I recommend computing them. in the TSP or Lua code.
    You can account for the phase shift in the computed source levels.
       
       -- compute the sine wave
        pts_per_cycle = 72
        wfrmTbl = {}
        for i=1, pts_per_cycle do
            wfrmTbl[i] = 1 * math.sin(i * 2 * math.pi/pts_per_cycle)
        end
        
        -- use some phase shift for smub sine wave
        wfrmTbl_smub = {}
        for i=1, pts_per_cycle do
            wfrmTbl_smub[i] = 1 * math.sin(i * 2 * math.pi/pts_per_cycle + math.rad(90))
        end
  • RE: 2636B: V pulse and transient current measurement

    You second task can be implemented with the asynchronous trigger model.

    Old forum showing the feature:
    asynchronous sampling from the source timing - capture transients

    I will upload a simple LabVIEW sample for loading scripts/functions and then making use of the functions from VIs.
     
  • RE: picoammeter 6485 firmware update problem

    Hello.
    I asked the factory AE to take a look at this.
    We suspect you have a serial number that has C01 firmware.
    If that is true, the unit cannot be down rev to Bxx firmware.

    Please provide screen shot of the *idn? response.
  • RE: IV measurements using arb waveform

    For PMU, there are three modes:
    two level pulsing (base and amplitude)
    SegArb
    FullArb

    The FullArb has no measurements;  it is sourcing only.
    The KPULSE utility is accessing the FullArb to play sine waves, etc.

    If you want IV measurements AND doing sine wave, you'd have to build it up from a piece wise segments in a SegArb sequence/waveform.

    Check out the pmu-dut-examples project and the pmu-segarb-complete test.
    It is using the PMU_SegArb_ExampleFull user module of PMU_examples_ulib.
    So long as you observe the rules for sequence defining, you can enter any SegArb that you desire on this.

    I suggest you contact your sales office/local applications engineer to get some detail on PMU.

     
  • RE: Simultaneous sine waves on SMU 2602 channels A & B

    Hello Fabrizio,

    Yes, this is very much possible with the 2600B SourceMeter.

    Sounds like you have our application note on ARB features of the 2600B SMUs
    Github code
    ARB Features for testing Ford EMC

    Either approach could work for generating sine wave from the SMUs.

    Suppose you use a list of 100 points to describe one cycle of the waveform.
    If you source those values at 10KHz rate, you’ll have the 100 Hz waveform.

    The max source rate is about 12K, so you’re approaching the limits of the speed.

    Keeping with the idea of 100 point list sweep:
    if trigger count is 100, then one cycle is sourced.
    if more than 100, the list is recycled.
    If trigger count is 0, then plays until you send abort command.

    If you don’t already, download the Test Script Builder application for running TSP code with the instrument.
    Then port it out to C# or Python, etc.

    Test Script Builder