-
RE: monitoring 6221 and 2182a using labview
Before using Labview or any other software, try just manual operation by pressing the Delta button on the 6221.
Reboot both. Press Delta button and accept the default parameter values. Press TRIG button when prompted.
Can that succeed?
Common pitfalls:
the RS-232 cable needs to be of type null modem. That is the type we ship with 6221/2182A.
On both the 2182A and the 6221, configure the RS-232 settings.
(19.2k BAUD, Flow is XonXoff, Terminator is CR).
After setting RS-232 settings on 6221, make GPIB or LAN the active interface. Otherwise you will see error +809.
-
RE: AFG310000 series not switching to sequence run mode
The SEQ does have separate license to have all features such as repeat, jump, etc.
Without the license, can load one waveform and continuous mode.
Obtain Trial License -
RE: Keithley 2410 turned into 2400 while being used in the production line.
What is the serial number?
-
RE: How do I get Kickstart for a DAQ6510
-
RE: Error Codes when trying to use TSP and error codes when trying to use python.
When running the above code, the following output occurs:***** KEITHLEY INSTRUMENTS INC.,MODEL 2636B,4597212,4.0.4 First status byte polling value: 0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 0 - 0b0 192 - 0b11000000 Test time: 3.765706600010162 Number of actual smua buffer pts: 1.51000e+02 Polling loop done, status byte: 128 Go get the data. After data downloaded, status byte: 128 Post reset, status byte value: 0
-
RE: Error Codes when trying to use TSP and error codes when trying to use python.
In Test Script Builder, look for the samples in folders labeled with 2600 such as KE26XXB_Example_Scripts.
In your post, you are attempting to use commands related to different products/different model numbers which have different command set from what the 2602B understands.
Here is some Python code for use with 2600B to carry out a voltage sweep and measure current.- import pyvisa as visa
- import time
- import sys
- import numpy as np
- import matplotlib.pyplot as plt
- resource_mgr = visa.ResourceManager()
- try:
- #get a session/handle for one resource
- #change the instrument_resource_string for YOUR INSTRUMENT
- #instrument_resource_string = "USB0::0x05E6::0x2602::4566591::INSTR"
- instrument_resource_string = "USB0::0x05E6::0x2636::4597212::INSTR"
- #instrument_resource_string = "TCPIP0::192.168.1.208::INSTR"
- my_instr = resource_mgr.open_resource(instrument_resource_string)
- except visa.VisaIOError as e:
- #did not connect.....
- print("Whoops, something went wrong")
- #print(e.args)
- #print(resource_mgr.last_status)
- print(resource_mgr.visalib.last_status)
- sys.exit(1)
- # set some parameters for our session
- my_instr.timeout = 25000 #timeout in msec
- #my_instr.write_termination = '\n'
- my_instr.VI_ATTR_TCPIP_KEEPALIVE = True
- my_instr.write("*IDN?\n")
- print("*****")
- print(my_instr.read())
- # set commands to setup a trigger model linear sweep
- num_sweep_pts = 151
- cmd_list = ["reset()",
- "errorqueue.clear()",
- "smua.source.func = smua.OUTPUT_DCVOLTS",
- "smua.source.limiti = 100e-3",
- "smua.source.rangev = 2",
- "smua.measure.nplc = 1",
- "smua.measure.delay = smua.DELAY_AUTO",
- "smua.measure.delayfactor = 1.0",
- "smua.measure.lowrangei = 100e-9",
- "smua.nvbuffer1.clear()",
- "smua.nvbuffer1.collecttimestamps = 1",
- "smua.nvbuffer2.clear()",
- "smua.nvbuffer2.collecttimestamps = 1",
- "smua.trigger.source.linearv(-1.0, 1.0, " + str(num_sweep_pts) + ")",
- "smua.trigger.source.limiti = 0.1",
- "smua.trigger.measure.action = smua.ENABLE",
- "smua.trigger.measure.iv(smua.nvbuffer1, smua.nvbuffer2)",
- "smua.trigger.endpulse.action = smua.SOURCE_HOLD",
- "smua.trigger.endsweep.action = smua.SOURCE_IDLE",
- "smua.trigger.count = " + str(num_sweep_pts),
- "smua.trigger.source.action = smua.ENABLE"]
- for cmd in cmd_list:
- my_instr.write(cmd)
- #commands to configure SRQ when SWEEPING bit goes 1 to 0 = sweep is finished
- # see Chapter 12 in 2600B Reference Manual
- cmd_list2 = ["status.reset()",
- "status.operation.enable = status.operation.SWEEPING",
- "status.operation.sweeping.enable = status.operation.sweeping.SMUA",
- "status.operation.sweeping.ptr = 0",
- "status.operation.sweeping.ntr = status.operation.sweeping.SMUA",
- "status.request_enable = status.OSB"]
- for cmd in cmd_list2:
- my_instr.write(cmd)
- #turn output on and run the sweep
- my_instr.write("smua.source.output = smua.OUTPUT_ON")
- t1 = time.perf_counter()
- # important to read stb once before starting trigger model
- print("First status byte polling value: " + str(int(my_instr.read_stb())))
- #start the trigger model operation
- my_instr.write("smua.trigger.initiate()")
- # ********************************************
- # setup a polling loop to detect Sweep is finished
- #repeat until the MSS (SRQ) bit is set
- still_running = True
- status_byte = 0
- debug = 1
- # attempts to my_instr.query("print(status.condition)") will be blocked by active trigger model task
- # instead use my_instr.read_stb() to obtain status.condition register
- while still_running:
- status_byte = int(my_instr.read_stb()) #read status byte (status.condition register)
- if debug: print(str(status_byte) + ' - ' + str(bin(status_byte)))
- if (status_byte & 64) == 64:
- still_running = False
- time.sleep(0.25) #250msec pause before asking again
- #turn output off
- my_instr.write("smua.source.output = smua.OUTPUT_OFF")
- t2 = time.perf_counter()
- print('Test time: {}'.format(t2 - t1))
- print("Number of actual smua buffer pts: " + str(my_instr.query('print(smua.nvbuffer1.n)')))
- print("Polling loop done, status byte: " + str(int(my_instr.read_stb())))
- print('Go get the data.')
- # retrieve data and graph it
- #ask for voltage and current; buffer2=voltage
- my_instr.write("printbuffer(1, smua.nvbuffer1.n, smua.nvbuffer2.readings,smua.nvbuffer1.readings)")
- raw_data = my_instr.read() #one long comma delimited string
- print("After data downloaded, status byte: " + str(int(my_instr.read_stb())))
- my_instr.write("status.reset()")
- print("Post reset, status byte value: " + str(int(my_instr.read_stb())))
- # split yields an array of strings alternating voltage, current, voltage, current
- raw_data_array = raw_data.split(",")
- volts = []
- current = []
- abs_current = []
- # step through the array of strings, step size 2
- # place the V and I into their own array of floats
- for i in range(0, len(raw_data_array),2):
- volts.append(float(raw_data_array[i]))
- current.append(float(raw_data_array[i+1]))
- abs_current.append(abs(float(raw_data_array[i+1]))) #absolute value of currents for log scale
- #plot the absolute value of current vs. voltage
- plt.plot(volts, abs_current, 'o-g')
- x_min = min(volts) * 1.1
- x_max = max(volts) * 1.1
- y_min = abs(min(abs_current)) / 10
- y_max = abs(max(abs_current)) * 10
- plt.axis([x_min,x_max,y_min,y_max])
- plt.yscale('log')
- plt.xscale('linear')
- plt.title('Sweep V, Measure I, 1MOhm')
- plt.margins(x=0.1, y=0.1)
- plt.grid(True)
- plt.show()
- save_image = 0
- if save_image:
- plt.savefig('alc_plot.png')
-
RE: Keithley 2410 turned into 2400 while being used in the production line.
Can you elaborate on what you mean by it turned into a 2400?
How does the instrument respond to the *IDN? command? The model number is part of the response string. -
RE: Solution for Current Pulse, Voltage Measure (150 usec, Digitizing mode) on 2461
Hi Gibbs,
First configure a digital line to give a trigger out when it receives the NOTIFY stimulus.
From the trigger blocks, you can add an entry to use the Notify event to cause the digital IO trigger.- --setup digital output to strobe when NOTIFY2 event occurs
- digio.line[1].mode = digio.MODE_TRIGGER_OUT
- trigger.digout[1].pulsewidth = 10e-6
- trigger.digout[1].stimulus = trigger.EVENT_NOTIFY2
- trigger.model.setblock(blockNumber, trigger.BLOCK_NOTIFY, trigger.EVENT_NOTIFY2) -- strobe the digital output
-
RE: External trigger of Keithley 2182A
The trigger lines in the trigger link are 5V, TTL digital lines.
The digital lines are agnostic about what they interface with, so long as they are TTL.
The same steps are required for the 2182A regardless of who’s current source is used.
Does the TDK have digital IO?
-
RE: Driver to control Keithley 3706A with visual basic application using ethernet cable
For basic communication over LAN, you can use raw sockets or you could obtain a VISA layer from Tek or NI, etc.
The IVI driver for 3706A contains the NI VISA runtime and leverages that for the communication.
There are some .NET examples distributed with the IVI driver.