RE: Multi-Channel Diode Test Options
Yes, it can do this with the DAQ6510 and a multiplexer card like 7701.
The instrument supports scans of channels where a measure function is assigned to each channel in the scan.
Scan supports the use of the Diode Test function where we force the selected amount of current and measure V using 10VDC range.
This use of Scan approach typically gives best throughput and easy to setup.
An alternate approach is to command a channel or channels to close. And then separately send commands to the internal DMM to config and perform measurements. Take note that closure of backplane relays to connect to the DMM is required as well as MUX channel closure.
In the case of model 7701 and Diode Test function, close "channel" relay 35 to connect card channels to internal DMM.
The first approach using scan takes care of backplane relays for you according to the assigned measure function.
See Multiple-Channel operation in the Reference Manual for DAQ6510. -
RE: Keithley 2601A IV pulse sweep and measurement
-
RE: Keithley 2601A IV pulse sweep and measurement
Do you have the Test Script Builder application?
If yes, at the TSP> prompt in the instrument console window you can have the full code listing of the factory KIPulse script printed out. You can then copy and paste it to a tsp file for editing.
You'll need to load the edited version to your own script/revised function names.- At the TSP> prompt type this:
- TSP>script.factory.scripts.KIPulse.list()
Alternately, there are example scripts distributed with Test Script Builder or we have them posted to our Github site.
2600A or 2600B Pulsed Sweep Examples -
RE: [k24xx series] Fetching data while outputting a list
For the courageous, here comes a sample for 2450 SMU.
First: I borrowed very heavily from the data streaming example for DMM6500 posted on our github:
Github: DMM6500 Streaming
Based on that template, I modified the lua file for SMU instead of digitizing DMM.
First, take note that 2450 SMU runs much slower than the 1MSample/sec digitizing DMM.
For the SMU, I put some functions that will ramp the voltage and take a current measurement at each voltage level.
There is also a repeat count parameter in case you want to cycle.
The revised Python and Lua code for streaming from 2450 is posted here:
GitHub: 2450 data streaming -
RE: [k24xx series] Fetching data while outputting a list
With 2400 or 2420, don’t think you’ll get there.
To have each measurement available on the PC essentially as they are available would require the PC controlling a loop. But as you observe, loops can have timing variation and changes with CPU speed, etc.
The use of trigger model allows better timing determinism, but as you observe, the instrument cannot serve up data and carry out the list sweep. It’s available only at completion of the sweep.
The 2450 and TSP could be pressed into using trigger blocks to have good timing control and run some code that prints the buffer data to output queue for pickup by a VISA Read. Keeping it coordinated will be a consideration. For example, if you do VISA read too frequently, you could get timeout errors that would need handled.
There is a nice app note and Python sample on our GitHub for data streaming from our DMM7510 of DMM6500. Similar approach can be applied to 2450, albeit at much slower rates than the DMMs.
App Note: (Appendix A: Python script for data streaming)
-
RE: 6221/2182A error-241“hardware missing”
Maybe obvious, but this error means the 6221 does not receive the expected response from 2182A on the rs-232.
Double check the rs-232 settings are same in both instruments.
Then set LAN or GPIB the active interface on 6221.
Press delta button on 6221 and accept defaults.
Error? -
RE: 2400 standard vs graphical RT voltage change does not exist with graphical
Hi Niko,
See if the attached document helps with how to access Live Update feature on the graphical SourceMeter such as 2450.
Andrea -
RE: 2461 Source meter
Again, looking into why these are not on tek.com manuals download site.
In meantime, it is attached here. -
RE: Keithley 6221/2182A in Delta mode
Can you post an additional image of the LabVIEW front panel for what values you are passing to the VIs? -
RE: Running measurement customizable scripts on Keithley 2636 via Python (PyVISA) - Speed problems
There is an example and code listing in the PDF I attached in earlier post.
This app note is good too:
App Note showing trigger model use - production test - fast throughput
Even for transfer curve, “sweeping” both gate and drain with trigger model task can give fast test speed.
If wanting a drain bias condition rather than a sweep, just make start equal to stop and number of steps same as the sweeping channel (gate). You’ll get an Ids at constant Vds for each Vgs value.
Sometimes you might want to do your own loop rather than use the trigger model sweep or list sweep. So what you have implemented is totally reasonable.
But if you judge it is taking too long to execute consider some revisions.
With the code you have now, each time your Python function is called, a lot of commands are sent to the SMUs. What’s in the Keithley_get()? Even more commands for the measurements?
All that is consuming time.
step 1: don’t execute setup commands needlessly. Same for reset. This costs time.
step 2: make a “apply configuration to SMU” function. Call it only when a setting is changing/prior to first run
step 3: make a run_it() function. Maybe clear buffers, output on, trigger.initiate(), <determine done>, printbuffers, output off - Next DUT
If you install these functions in the runtime memory of the SMUs, then the amount of characters across the bus between Python and the instrument can be vastly reduced and your speed further improved. You just need to VISA write the name of your functions to execute the code on the instrument. Big speed improvement!
I also encourage you to make use of tsp-link so that your Python code only needs connection to one instrument. Over tsp-link you will have node[2].smua and node[2].smub rather than a second VISA session to an instrument.
Last on buffer size: how many samples do you need? In the rare case that nvbuffers are not large enough, you can allocate runtime memory to be an even larger buffer. If you are getting past 100K samples to be transferred, take a look at binary xfer to reduce bus traffic for the data to PC xfer to again reduce time to do the task.