RE: Entry Point Error in GPIB diagnoistic
Check your gpib-32.dll properties. Is it from NI or Keithley or possibly CEC?
My guess is that something installed the NI driver for GPIB. And it over-wrote the required version of that DLL needed by the KPCI-488A card.
Can you reinstall the driver for KPCI-488A? -
RE: DAQ6510 Display when using rear terminals under GPIB control
Did you close the channel?
For the reading that is placed in the buffer, what channel tag does it have?
My .NET code from prior post does cause the front panel to update the watch channel measured value.
On the instrument, in the event log settings, you can enable command logging. This will let you see exactly what the IVI driver commands are sending to the DAQ.
If still not working, check your firmware level. I'm at 1.7.12b
-
RE: Keithley 2614B : Unable to detect my SMU in Test script builder
Check that you have the most recent version of Test Script Builder (J05).
Also check the firmware on your 2612B (3.4.2 is most recent).
Test Script Builder is making use of NI-VISA for communication with the instrument.
From error screen, it is not clear what the issue is.
If you have NI-MAX installed, try verifying basic VISA IO from that.
If you still need help, tell us the VISA resource name that you use for the 2612B.
-
RE: 2461 SourceMeter pulse train clipping at 36 V short of 105 V spec
What is the state of interlock?
If interlock not satisfied, the max voltage is clipped/only a portion of 100V range can be used. -
RE: DMM7510 ignores external trigger event
Closing the loop on this one.
Suggest updating firmware to latest (1.7.12b) ; you are at 1.7.7b.
Even so, when using the integrating A/D, we are able to duplicate the issue with DMM7510 but not with DMM6500.
A bug has been logged for DMM7510 behavior.
Workarounds if using DMM7510:
First: your trigger model operation is essentially sampling until told to stop; You could get to essentially same outcome by making use of abort() command to stop an infinite count acquisition into a recycled finite sized buffer.
Second: your trigger model operation will work and *TRG can be detected/used to stop it if the digitizing A/D is used instead of the integrating A/D.
Here is a code sample:my_instr.write("reset()") my_instr.write("errorqueue.clear()") #comment out the use of integrating A/D #my_instr.write("dmm.measure.func = dmm.FUNC_DC_VOLTAGE") #my_instr.write("dmm.measure.range = 0.1") #use the digitizing A/D my_instr.write("dmm.digitize.func = dmm.FUNC_DIGITIZE_VOLTAGE") my_instr.write("dmm.digitize.range = 0.1") my_instr.write("dmm.digitize.samplerate = 10e3") my_instr.write("dmm.digitize.aperture = dmm.APERTURE_AUTO") my_instr.write("measure_buffer = buffer.make(12)") my_instr.write("measure_buffer.fillmode = buffer.FILL_CONTINUOUS") my_instr.write("trigger.model.setblock(1, trigger.BLOCK_BUFFER_CLEAR, measure_buffer)") my_instr.write("trigger.model.setblock(2, trigger.BLOCK_MEASURE_DIGITIZE, measure_buffer, trigger.COUNT_INFINITE)") my_instr.write("trigger.model.setblock(3, trigger.BLOCK_WAIT, trigger.EVENT_COMMAND, \ trigger.CLEAR_ENTER, \ trigger.WAIT_OR, \ trigger.EVENT_DISPLAY)") my_instr.write("trigger.model.setblock(4, trigger.BLOCK_MEASURE_DIGITIZE, measure_buffer, trigger.COUNT_STOP)")
-
RE: KTH-2651A Minimum Operating Voltage as Load
Stands to reason that the SMU that is sinking the current needs to be lower potential than the external voltage source that the SMU is loading.
If need be, the SMU will transition to quadrant 3 (negative current and negative voltage) to allow the source current setpoint to be achieved.
See if the attached PDF helps.
What are you proposing to connect the 2651A to? -
RE: KTH-2651A Minimum Operating Voltage as Load
Can I ask: what do you need the operating voltage to be?
I've uploaded a block diagram level of the SMU.
For your question about using it as a load, you will operate it as a current source with a voltage compliance limit.
To operate in quadrant 4 as a load, you would have the SMU source negative current.
The V-meter provides input to the voltage compliance limiting.
-
RE: TSB Config Panel wont' install 3706A USB driver
Hello,
That is very perplexing.
The USB0:: resource string is what I expect.
Does the front panel display the REM annunciator to indicated the remote connection has been made?
Also expect TSB will issue a reset(). For unknown reason, it is not successfully receiving a response from the 3706A.
Try a different USB A-B cable. Also try different USB port on the computer if you have one.
With all Keithley related software, you should also have a utility program: Keithley Communicator.
Try: close TSB and reboot 3706A. Invoke Communicator and open the same USB0:: resource.
You and enter commands in the bottom frame, then hit the run button to send the command.
If you send something that sends back a response, press F2 or menu Command>Read Instrument Data.
Can you send *idn? command and read the response back?
What about sending reset() command. And errors displayed on the instrument front panel?
Attached are some screen shots.
-
RE: TSB Config Panel wont' install 3706A USB driver
In Test Script Builder (TSB), the 'Open Instrument' icon in the Instrument Console should be enumerating and displaying the detected USB0:: type of VISA resource string for your model 3706A.
Do you have that? If yes, select that and see if it can successfully connect.
The Wizard that you mention sounds like the Keithley Configuration Panel. There is no requirement to do anything in the Config Panel to have TSB be able to see the USB0:: resources or to subsequently use the instrument from Python or similar. The config panel allows you to associate a VISA resource with an IVI driver and also to assign alias to it to use when opening communication sessions. The NI MAX software tool covers these same optional configuration aspects. -
Amps Measure Function Protective Fuse for DMM4020
The 200 mA input is protected by a fuse (F2) rated at 440 mA, 1000 V (fast blow).
Dimension: 0.406 IN diameter X 1.375 IN length
Higher ranges are protected by fuse (F1) rated 11Amps, 1000V (fast blow).Dimension: 0.406 IN diameter X 1.5 IN length
These fuses are fairly popular in many different DMMs with amp measure.
Consider Fluke:
FUSE-440MA/1000VB1
FUSE-11A/1000V B1
Littlefuse FLU.440 and FLU011
Ferraz Shawmut FSM44/100 and FSM11