• RE: TekTalk Search Not Working

    Need to set the dropdown to Forums:

  • RE: How to capture screen via USB on MSO44

    Check out the free TekScope Utility.
    TekScope Utility
  • RE: eScope not display scope screen but a broken image icon instead

    Escope can be finicky with the older TDS3000B series scopes.  If you are unable to get a browser combo to work consider looking at TekScope Utility. 
  • RE: Simultaneous Source Output and Digital IO

    Attached is a document that describes how to make use of the TSP functions and scripts from LabVIEW.
    The LV driver for 2600x has a Load Script VI.
  • RE: Simultaneous Source Output and Digital IO

    Great progress.  Let me give you some trigger model sample code to illustrate the very tight timing you can achieve.

    Attached is a scope capture.
    The digital IO have pull up resistors so will be at 5V or logic HI by default.  Looks like you want a rising edge trigger for your DAC board, so you've been using a mode command to change the DIO state.  The scope capture starts at that occurrence.

    Subsequently, when the trigger model task is started, the DIO goes to 5V and the SMU goes to 10V.
    function config_smu(v_src, iLimit)
        smua.source.func = smua.OUTPUT_DCVOLTS
        smua.source.limiti = iLimit
        smua.source.rangev = v_src
        smua.source.levelv = 0   -- this is the resting bias level as soon as output is turned on
        smua.measure.nplc = 1
        smua.measure.delay = smua.DELAY_AUTO
        smua.measure.delayfactor = 1.0
    end  -- function
    
    function config_dio()
        digio.trigger[1].mode = digio.TRIG_RISINGM
        digio.trigger[1].stimulus = smua.trigger.ARMED_EVENT_ID
        digio.trigger[1].pulsewidth = 1.000000
    end  -- function
    
    function config_trigger_model()
        smua.trigger.arm.count = 1
        smua.trigger.arm.stimulus = 0
        smua.trigger.source.stimulus = smua.trigger.ARMED_EVENT_ID
        smua.trigger.source.action = 1
        --smua.trigger.source.linearv(10.000000, 10.000000, 1)
        smua.trigger.source.listv({10})   -- single point list sweep is another way
        smua.trigger.source.limiti = 0.1
        --smua.source.delay = 1.000000
        smua.trigger.autoclear = smua.ENABLE
        smua.trigger.count = 1   -- set this equal to number of sweep or list points
        smua.trigger.endpulse.action = smua.SOURCE_HOLD
        smua.trigger.endsweep.action = smua.SOURCE_HOLD
    end  -- function
    ​​​​​​​
    -- **************************************************
    -- use the functions
    reset()
    errorqueue.clear()
    config_smu(10, 0.1)
    config_dio()
    config_trigger_model()
    smua.source.output = smua.OUTPUT_ON
    
    delay(0.5)  -- not needed....just allows us to see the dio setup of getting to LO state
    smua.trigger.initiate()
    delay(1.25)  -- not needed...just delay a bit before setting the SMU source level back to zero
    smua.source.levelv = 0

     
  • RE: Simultaneous Source Output and Digital IO

    Hi Matt,
    Those are great observations.
    For either approach, as your "resting state" can you have the SMU output (the blue light) on, but have it forcing 0V?
    Then when ready for testing, command the digital IO and the source level to change to a new level.

    If I measure the time required to turn the smua.source.output on and off wiht the source level set to 3V, I'm getting about 8 msec for that operation.
    When output is off, the SMU is still actively sourcing and limiting according to the output off mode and settings.
    When output is then turned on, the new settings have to be applied, and you are seeing some overhead associated with that.

    In contrast, if the output is already on but at 0V, then changing the source level to a different level requires much less time.

    Combining this info about the source output and my prior suggestion about using a function to reduce bus traffic, I'm seeing less then 40usec from the DIO changing state and the start of the SMU source level changing.

    Some TSP code to illustrate the concept:
    function config_smu(v_src, iLimit)
        smua.source.func = smua.OUTPUT_DCVOLTS
        smua.source.limiti = iLimit
        smua.source.rangev = v_src
        smua.source.levelv = 0   -- this is the resting bias level as soon as output is turned on
        smua.measure.nplc = 1
        smua.measure.delay = smua.DELAY_AUTO
        smua.measure.delayfactor = 1.0
    end  -- function
    
    function speed_test(dio_line, v_src)
        -- to run the configured setup
        digio.writebit(dio_line, 0)  -- make digio low state
         smua.source.levelv = v_src
        digio.writebit(dio_line, 1)
    end  -- function
    
    -- test our function
    reset()
    errorqueue.clear()
    digio_line = 1
    config_smu(2.5, 0.1)  -- config for 2.5V and 100mA
    smua.source.output = smua.OUTPUT_ON
    for i = 1, 10 do
         --speed_test(v_src)
        speed_test(digio_line, 2.5)
        delay(100e-6)
        smua.source.levelv = 0
       
    end
    smua.source.output = smua.OUTPUT_OFF
    

    A screen shot from the scope is attached.
  • RE: Simultaneous Source Output and Digital IO

    For sure, we can get the timing of the DIO and the sourcing more synchronized.
    Is the 2601B doing anything else?  Any measuring?
    Does the source level stay applied until you send new commands?

    To have the best synchronization will require use of the trigger model.
    Both the DIO output trigger and the source stimulus can be setup to use the same start event (or software assert).
    With this approach you can get the synchronization to sub microsecond.
    You can even deliberately time skew them if you wanted to trigger and then source a known amount of time later.

    Before re-architecting the code for trigger model, you could most likely realize some time improvement by defining a function in the runtime memory of the 2601B.  Your LabVIEW code would need to use a VISA level write to call your function.  The function would issue the DIO, output on, etc like what you are doing now.  The time savings comes from requiring just a single bus command between LabVIEW and the 2601B.
    The function execution would still sequentially carry out the lines of code.  But the command parsing from the bus is reduced to just once.
  • RE: DAQ6510 problem changing resistance measuring range using SCPI

    :SENS:FUNC "RES", (@101)
    :SENS:RES:RANG 10000, (@101)
    :ROUT:CLOS (@101)
    :READ?
  • RE: MSO24

    You can connect over LAN or USB.

    Over LAN, you can use something like TightVNC to interact with the scope.

    Other software approaches:
    KickStart scope application
    TekScope PC
    TekScope Utility (free)
    Python drivers
    more
  • RE: Keithley 2750 DAQ DMM w/ 7702 MUX DAQ card. Detecting shortcuts in multiple pairs.

    Very nice implementation.  Thanks for sharing!